Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Routine Maintenance 700 Series ICP-OES Systems
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP OES Systems
  • Product Name :  700 Series ICP-OES
  • Task :  Best Practices Maintenance
Related Articles
  • How to perform a Torch Alignment Calibration on the Agilent 700 Series ICP-OES

    This Information Applies To: Agilent 700 Series ICP-OES instruments Issue: It is important that a torch alignment is performed to ensure that the optics are optimized and are focused on the analytical zone of the plasma. This alignment procedure will ensure maximum sensitivity and signal-to-noise. It is recommended this procedure is carried out regularly, at least once per week, to ensure optimal performance. The Agilent standard test solution (p/n 6610030100 ) can be used as it contains...
  • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) and Agilent 700 Series ICP-OES (710/715/720/725/730/735) Issue Loss or change of sensitivity in ICP-OES can occur under normal operating practices. This discussion identifies factors that can contribute to sensitivity changes including calibration standards , memory effects , and the sample introduction system . Sample introduction components include: Autosampler probe Switching valve Pump tubing Nebulizer...
  • Preparing the Instrument Calibration Solution for the Agilent ICP-OES System

    This Information Applies To: All ICP-OES: 5000 (5100/5110/5800/5900) and 700 (710/715/720/725/730/735) series Issue The instrument calibration solution is used when performing the ICP-OES Performance Test and the Wavelength Calibration. The instrument calibration solution can be prepared from 1000 ppm stock solutions ( Table 1) according to the calculations described in this article. Alternatively, you can purchase the Agilent ready to use wavelength calibration standard ( Table 1 ). Instrument...
  • ICP-OES Torch Overheating

    This Information Applies To: Agilent ICP-OES Systems (700 Series, 5000 series) Issue If any yellow or orange coloration on any part of the torch is observed (not due to the sample being presented or recently run) or if the torch appears to be visibly melting, the torch is overheating. Figure 1. Example of damaged torch Steps to Follow Caution : To prevent damage to the instrument, do not use the torch if it is damaged Warning : The torch and torch compartment...
  • AVS 6/7 Maintenance Checklist

    This Information Applies To: Agilent 5000 Series ICP-OES, AVS 6/7 Issue The AVS 6 and AVS 7 require routine maintenance. This article provides a maintenance checklist. Figure 1. 1 AVS 7 Warning : Always rinse all tubing and the valve before disconnecting the tubing from the switching valve to avoid contact with potentially dangerous chemicals. Appropriate personal protective equipment (PPE) should be worn when handling components contaminated with chemicals. ...

Routine Maintenance 700 Series ICP-OES Systems

Answer

This Information Applies To: Agilent 700 Series ICP-OES Systems


Issue 

The 700 Series ICP-OES requires regular maintenance, and routine maintenance tasks are able to be performed by the instrument operator. 

 Tip  Any maintenance procedures not specifically mentioned in this article should only be performed by Agilent-trained, Agilent-qualified, or Agilent-authorized Field Service Engineers.

 

 

 Warning  Warning: Eye Hazard Danger to eyes. The plasma is an intense light source. Direct viewing of the light source will cause eye damage. Operators and other unauthorized personnel must NEVER remove the main covers or disable the safety interlock system.
 Warning  Warning: Shock Hazard. This instrument contains electrical circuits, devices, and components operating at dangerous voltages. Contact with these circuits, devices, and components can result in death, cause serious injury, or painful electric shock.
 Warning  Warning: RF Hazard and Hot surfaces. The plasma radiates dangerous levels of radio frequency (RF) energy. Exposure to this RF energy can cause severe skin damage and cataracts of the eyes. Close contact with the operating plasma can result in severe heat burns to the skin, and an electrical discharge that can jump a considerable distance and may cause death, severe electric shock, or subsurface skin burns.

The following parts of the ICP-OES require routine maintenance. Detailed maintenance instructions are included in the Help menu of the ICP Expert II software.

Daily

  • Inspect nebulizer and torch
  • Check the water level in the Argon Saturator Accessory (ASA) before every use (if applicable).
  • Empty drain vessel
  • Inspect the pump tubing and replace if it has lost its elasticity. Unclip the pump tubes when the pump is not in use.

Weekly

  • Clean the torch.
  • Clean the cone (axial) or snout (radial)
  • Inspect the bonnet (radial).
  • Clean the spray chamber.
  • Clean the nebulizer.

Monthly

  • Clean the cooling air intake filter (behind chimney).
  • Inspect the induction coil. Contact your local Agilent office or representative if maintenance is required.  
  • Remove and clean the water filter on the back of the instrument.
  • Check the water level in the water cooler (see the manual supplied with your water cooler for details).
  • Check/clean the heat exchanger (radiator) on the cooling system to remove any build-up of dust and dirt.
  • Periodically, drain the coolant from the cooling system and then refill/ treat with an appropriate algaecide (as recommended by the manufacturer).
  • Perform a wavelength calibration.
  • Inspect the external gas supply system for leaks including the tubing connected to the instrument. Replace any damaged, leaking, or worn components.

 

 

 Tip 
Learn how to effectively maintain and troubleshoot your Agilent 700 Series ICP-OES System:
Agilent 700 Series ICP-OES Maintenance Tips
 
 