Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
Greetings Agilent Community,
As with many of the errors I encounter with my system, these two happened after I foolishly left the lab to pursue a life outside the world of LCMS.
(1) From the instrument log, the first error read "Failed to run acquisition…
When attempting to submit a sequence in EZChrom, it fails to submit and returns this error message:
This happens each time a sequence is submitted. Any help would be appreciated!
Hi Agilent community.
I'm a LC-MS equipment user and I had a problem with the MassHunter Acquisition B 08.01 software.
When I try to open the Acquisition software, appears the software logo but immediately appears an error dialog box. This error dialog…
Hi everyone,
We have a 6546 QTOF, the following error keeps popping up during acquisition; Failed to start logbook server. A shared violation has occurred. How can I resolve this?. I'll appreciate your help. Thanks
How to Access Spectroscopy Database Administrator Software (SDA)
Multiple Results
This Information Applies To: Agilent Spectroscopy Database Administrator Software
Issue
The SDA application can only be accessed by a Windows user who is a member of the Windows Administrators group and Windows SDAAdministrators group.
Steps...
This Information Applies To: AriaMx real-time qPCR system
Issue
The instrument fails to reach the required temperature during a run.
Resolution
Check if the temperature of the environment around the instrument is too low resulting in the instrument...
This Information Applies To : Agilent ELSDs G4260A, G4260B, G4261A, G4621B, G7102A, G7102B
Content :
If the evaporator tube of ELSD becomes contaminated with nonvolatile material resulting in poor chromatography, please use the procedure in this...
This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
The oil should be replaced at least once every six months, or when the oil color is indicates it needs to be changed . The foreline pump should be checked monthly. This article...
This Information Applies To: Agilent ICP-MS MassHunter versions 4.4 - 4.6 and 5.x.
Issue
After performing an autotune, the batch tune report will not generate or print.
Background
ICP-MS MassHunter uses an application called ICP-MS Data Analysis...
This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems equipped with a gate valve
Issue
Installing the Ion Transfer Capillary after cleaning or replacement in Agilent LC/MS instruments with a gate valve.
Required Items...
This Information Applies To: Agilent OpenLab CDS 2.x
Issue
This article outlines the procedure on how to quantify uncalibrated compounds in OpenLab CDS 2.x. using reference compounds in the processing method compound table.
Background
Uncalibrated...
This Information Applies To: Agilent OpenLab CDS 2.x
Issue
This article outlines the procedure on how to quantify the uncalibrated compounds in OpenLab CDS 2.x when the analytical method requires a fixed response factor.
Background
Uncalibrated...
This Information Applies To: Agilent OpenLab ChemStation Edition Software (C.01.xx)
Issue
When your Agilent OpenLab ChemStation Edition software fails to launch with errors " Login failed with the following message: Configuration system failed to...
This Information Applies To: Agilent OpenLab ChemStation Edition Software (C.01.xx)
Issue
Agilent OpenLab ChemStation Edition software fails to launch with error containing the words " Autostart Macro failed ”. This article will provide you some...