Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles Cleaning Recommendations for Agilent ELSDs
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  9 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems
  • Component :  LC Detectors
  • Product Name :  1260 Infinity II LC System 1260 Infinity LC System 1260 Infinity II Prime LC System 1290 Infinity II LC System 1290 Infinity LC System
  • Task :  Best Practices Maintenance
Related Articles
  • How to Prepare your HPLC for a Long Time Shut Down

    This Information Applies To: Agilent Liquid Chromatography (LC) systems Issue: Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column. Steps to follow: Store the column in an appropriate solvent (check column information sheet) When you remove the column, add a column union so the detector gets flushed as well in the following steps. For a system in r eversed phase -> Purge all lines with hot HPLC-grade...
  • How to Submit a Shutdown Method in OpenLab 2.x

    Multiple Results
    This Information Applies To: Agilent OpenLab CDS 2.x controlling Agilent LC Systems Issue This article outlines an instruction on how to submit a shutdown method on OpenLab CDS 2.3 or above. Background Submit a shutdown run to the run queue...
    • Found In:
    • Articles
    • ,
    • Articles
  • How to Clean the Capillary Array Window on Agilent Parallel Capillary Electrophoresis Instruments

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, and ZAG DNA Analyzer Issue: Dirt or dust on the capillary array window can cause baseline shifts or elevation of the baseline. Figure 1 shows an example electropherogram with baseline shifts. It is sometimes possible to see the dirt or dust in the View Capillary Positions screen in the Agilent ProSize data analysis software. This screen can be opened by clicking Analysis>View Capillary Positions in ProSize. An example...
  • Loss of Peak Resolution in GC

    This Information Applies To: Agilent GC Systems Issue Loss of peak resolution can be a combination of decrease in separation and increased peak width. Loss of resolution can impact on analytical reproducibility and accuracy. Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument in accordance with the recommended maintenance schedule will reduce the incidence of chromatography problems. Background Loss of resolution...
  • General Recommendations for Using an Agilent RID

    This Information Applies To: Agilent Refractive Index Detectors, G1362A and G7162A/B. Issue: The RID is sensitive to its environment and can experience stability problems when exposed to contamination, vibrations, mechanical shocks, or changes in temperature or pressure. Resolution: Considerations on the general set-up: To avoid cross-talk between different solvent channels, preferably use an isocratic pump when operating a RID. If it is not possible to use an isocratic pump, consider...

Cleaning Recommendations for Agilent ELSDs

Answer

This Information Applies To: Agilent ELSDs G4260A, G4260B, G4261A, G4621B, G7102A, G7102B


Content:

If the evaporator tube of ELSD becomes contaminated with non-volatile material resulting in poor chromatography, please use the steps below to troubleshoot and clean the ELSD.


Steps to Follow:

It is recommended that the instrument is initially washed with a solvent suitable to clean the contamination. 

  Caution
Caution: Do not use solvents that contain additives when performing the cleaning procedure.

 

 Caution 
Caution: Please read all relevant MSDS for the solvents being used, and follow all safety guidance including storage, handling, ventilation and extraction requirements, and the use of appropriate PPE.

One of the “cleaning” solvents recommended is Acetone/Water (50/50) (based on your application, select the appropriate solvent to clean the contamination).

Cleaning Procedure: 

  1. Set Evaporator temperature to 40 ?.
  2. Set the Nebulizer temperature to 40 ?.
  3. Set gas flow to 2.5 SLM.
  4. Connect the ELSD directly to the HPLC pump (bypassing the ALS and column).
  5. Pump the “cleaning” solvent into the instrument at 1 or 2 mL/min (while in the RUN mode).
  6. Perform this procedure overnight or for a minimum of 4 hours. Make sure that there is the required volume of cleaning solvent for the flushing time. 
  7. During the cleaning procedure, the ELSD response may show a full-scale deflection. This indicates that contamination is being flushed from the detector.
  8. Following the cleaning procedure, reconnect the column and ALS and inject a sample of known response.
  9. Repeat the wash procedure with a different solvent, or for a longer period if the ELSD signal is not improved.

Preventative Maintenance:

Depending on use, it is recommended to clean the evaporator tube at least once a week or every 40 hours.

The build-up of contamination can be identified early by monitoring the baseline noise over time, or by checking the ELSD offset reading.

To check the ELSD offset value;

  1. Switch off the ELSD.
  2. Switch the detector back on and place it into RUN mode.
  3. Record the ELSD reading from the front panel -this is the OFFSET value
  4. Offset values above 140 mV indicate contamination and it is recommended the ELSD be cleaned.

It is also recommended to clean the unit following the use of buffers, or where the sample matrix contains impurities (i.e. "dirty" sample).

If cleaning the unit does not cure the problems then contact Agilent Technologies for further assistance.

 

 Tip
Learn how to effectively maintain your Agilent ELSD:
ELSD User Manual on Agilent.com
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help