This Information Applies To: Agilent ICP-MS MassHunter versions 4.4 - 4.6 and 5.x.
Issue
After performing an autotune, the batch tune report will not generate or print.
Background
ICP-MS MassHunter uses an application called ICP-MS Data Analysis File Messaging Service - Task Queue Viewer to generate reports (such as batch tune reports, or batch sample reports) from *.acrt template files.
If a task in the Task Queue Viewer is stuck in Processing status, reports will cease to generate.
Powering off the computer off, then back on, and retrying to generate the report will often resolve the issue. If the issue remains, it is necessary to clear all tasks from the Task Queue Viewer and troubleshoot why the reports will not generate.
Steps to Follow
Tips: Before continuing, close the MassHunter software, power off the computer, then power the computer back on and retry generating the report. If the issue is fixed, no further action is required. |
Figure 1. ICP-MS Data Analysis - File Messaging Service icon
Tips: Hold Shift on the keyboard and left mouse click each line to highlight multiple lines at once.
|
Figure 3. Agilent MassHunter Workstation - Task Queue Viewer
1. Tasks with status 'Done' or 'Unknown' highlighted, 2. Delete Tasks button
Caution: Once deleted, these waiting tasks will not generate a report. It will be necessary to regenerate the missing reports once the issue is resolved.
|
Figure 4. Agilent MassHunter Workstation - Task Queue Viewer
1. Tasks with status 'Waiting' highlighted 2. Delete Tasks button
Figure 6. Deleting tasks
1. Tasks with status 'Processing' highlighted 2. Delete Tasks button 3. Note: the stop button is no longer enabled
Figure 8. Windows Task Manager
1. Processes Tab, 2. IcpReportingPrc application, 3. End task
Figure 9. Batch open in ICP-MS MassHunter Software
1. Tune Modes Tab (MassHunter 5.x) 2. Tune Tab (MassHunter 4.x)
Figure 10. Print Current Signal from Tune Modes (or Tune) Tab
Figure 11. ICP-MS Data Analysis - File Messaging Service icon
Figure 13. Settings page
1. Reporting tab, 2. Printer check box not selected, 3. Close button
Figure 14. Task Queue Viewer
1. Tasks with status "Unknow" highlighted, 2. Delete Tasks button
Figure 16. Generate Tune Report from Tune Modes (or Tune) Tab
Figure 17. Selecting a Report Template File
1. Template selector icon 2. Example Tune Report Template File selected 3. Open button 4. Generate button