Hello ivynotterpek ,
I'm not aware of any way to do this with just one compound/cal curve. One solution would be to make duplicate compounds and assign one compound the low level standards and the other the high level standards. You can utilize different…
Thank you, unfortunately we tried to set a collection window much larger than the peak duration, but the peak got split again. And in the version of the software we are using no advanced settings allowing to collect different peaks into the same vial…
Hi CMZainea,
Thank you for your feedback. The leak test for the 6 ports valve was passing, so no leaks detected at the valve. Any other leaks were detected from the visible piping and connection from the back panel of the 7697A going to the 6 ways valve…
The two-way splitter without makeup G3181B can be used as a splitter or a combiner to connect two inlets to one detector. Discuss this part with your local sales rep and they can help you get a quote for this part.
https://www.agilent.com/cs/library…
The Digest of upcoming webinars has been updated as of February 20.
Notable additions include:
BioPharma (1)
GC & GCMS (2)
Pharma (1)
Software & Informatics (21)
Check Out:
Tips and Tricks Workshop on ICP-OES: Smarter Method Development…
This Information Applies To: Agilent ICP Expert software.
Issue
The software version is often asked and is a valuable information: to see if any updates are needed, if certain options are available etc.
Steps to Follow
Launch ICP Expert...
This Information Applies To: Agilent OpenLab CDS
Issue
This article provides step-by-step instructions to calculate peak-to-valley ratio in OpenLab CDS.
Background
The peak-to-valley ratio is calculated to indicate the quality of peak separation...
This Information Applies To: Agilent 6500 Series Q-TOF LC/MS systems, MassHunter Workstation 10.1 or below and MassHunter Workstation 11.0 .
Issue
Fully reset the Q-TOF LC/MS system via a power cycle without venting. For users who experience slow...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, and Oligo Pro II systems
Issue
The volume of sample used by Agilent parallel capillary electrophoresis systems during a run.
Resolution
The...
This Information Applies To: Agilent 2100 Expert software, 2100 Expert Security Pack, 2100 Bioanalyzer
Issue
When opening the 2100 Expert software and using the Security Pack, the error message " Start of launcher application failed - The requested...
This Information Applies To: Agilent 6890, 6850, 7890, 8890, 8860, Intuvo 9000 GC systems.
Issue
The exchange of the ignitor in a Flame Photometric Detector (FPD) may be required in the following scenarios:
Preventive or periodic maintenance...
This Information Applies To: Agilent 1290 Infinity Quaternary Pump G4204A, 1290 Infinity II Flexible Pumps G7104A/C, and G7131A/C.
Issue
If the flexible/quaternary pump does not generate flow or build up pressure, it could be due to a stuck inlet...
This Information Applies To: Agilent 5200, 5300 and 5400 Fragment Analyzer systems
Issue
Separation times for the Fragment Analyzer for each array format and kit.
Resolution
Caution : The times in this article are the length of time...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, Oligo Pro II, ProteoAnalyzer systems
Issue
Guidance for installing the Agilent parallel capillary electrophoresis instrument controller and/or...
This Information Applies To: OpenLab CDS (version 2.x) system
Issue:
This article explains how to export OpenLab CDS 2.x data as Archived file.
For copying OpenLab CDS 2.x data to another Projects or another OpenLab CDS systems, you can use data...