This Information Applies To: Agilent 6890, 6850, 7890, 8890, 8860, Intuvo 9000 GC systems.
Issue
The exchange of the ignitor in a Flame Photometric Detector (FPD) may be required in the following scenarios:
Items Required
Steps to Follow
|
Caution: Wear clean powder-free gloves when working on FPD components in order to prevent contamination.
|
|
Warning: Cool the FPD and shut off the high voltage before any proceeding. The detector may be hot enough to cause burns. If the detector is hot, wear heat-resistant gloves to protect your hands.
|
There are two FPD detector types available and the procedure is slightly different:
Standard FPD
Figure 1. View on the standard FPD and the collet screw.
Figure 2. View on the standard FPD
1. O-ring, 2. Ignitor
Figure 3. View on the block
1. O-ring, 2. Spacer
FPD+
Figure 4. View on the FPD+ and the collet screw
Tips: Use a 5/16” nut driver, a wrench will not fit.
|
Learn how to effectively maintain your Agilent GC system:
Agilent 8890 GC with OpenLab CDS eLearning path available from Agilent education |