This Information Applies To: Agilent 2100 Expert software, 2100 Expert Security Pack, 2100 Bioanalyzer
Issue
When opening the 2100 Expert software and using the Security Pack, the error message "Start of launcher application failed - The requested operation requires elevation" appears.
Background
This error appears for both to users and administrators, making it impossible to log on to the 2100 Expert software (Figure 1).
Resolution
. To ensure proper functionality of the Expert Launcher, set up the correct Expert Launcher properties:
Figure 3. ExpertLauncher Compatibility for all users menu. The Run this program as an administrator check is highlighted with a red rectangle.
If the problem persists, contact your local Agilent support representative.
Learn how to effectively maintain and operate your Agilent 2100 Bioanalyzer system: 2100 Expert Software User's Guide on Agilent.com Agilent 2100 Bioanalyzer System Maintenance and Troubleshooting Guide on Agilent.com |
For Research Use Only. Not for use in diagnostic procedures.