This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, Oligo Pro II, ProteoAnalyzer systems
Issue
Guidance for installing the Agilent parallel capillary electrophoresis instrument controller and/or ProSize data analysis software. You can install the software on a different PC or to upgrade to the latest software version.
Resolution
The controller software and ProSize analysis software for each parallel capillary electrophoresis instrument can be downloaded from the following webpages on the Agilent website:
For each software download page, the ProSize software section is at the top, but if you scroll down, you will see the link to download the instrument controller software. Read me files, which detail each software’s system requirements can be downloaded from the same webpage.
Caution: If the software is installed on a computer, which was not supplied by Agilent, Agilent cannot guarantee the performance of the instrument with the PC. It is best practice to back up all data, custom methods, and custom configurations before upgrading the software or changing the PC. Storage locations can be found in the following instructions.
|
Installing ProSize
If you are installing software only for data analysis purposes, and not to control the instrument, install the ProSize data analysis software. You do not need to install the instrument controller software.
If you are installing the software on a different PC:
Tip: If a previous software version is already installed and you are upgrading the software, it is not necessary to copy over the previously mentioned files. However, before the software upgrade, it is best practice to back up the files, along with all data, custom methods, and custom configurations. You do not need to uninstall the old version when upgrading software. The new version should just overwrite the old software version. |
Installing the Instrument Controller Software
If you are installing the software on a different PC:
1. From the folder (where instrument is the name of your instrument type) C:\Agilent TechnologiesI\Instrument or C:\AATII\Instrument copy the following files:
2. Copy any custom method files from the folder C:\AATI or Agilent Technologies\Methods.
Tip: If a previous software version is already installed and you are upgrading the software, it is not necessary to copy over the previously mentioned files. However, before the software upgrade, it is best practice to back up the files, along with all data, custom methods, and custom configurations. You do not need to uninstall the old version when upgrading software. The new version should just overwrite the old software version. |
Tip: Installing the controller software on a new PC will not result in an instrument crash. The stage location info is stored on the stage board.
|
Learn how to effectively operate your Agilent parallel capillary electrophoresis system: 5200, 5300, and 5400 Fragment Analyzer system manual on Agilent.com Femto Pulse user manual on Agilent.com ZAG DNA Analyzer system manual on Agilent.com Oligo Pro II user manual on Agilent.com ProteoAnalyzer system manual on Agilent.com ProSize data analysis software user manual on Agilent.com |
For Research Use Only. Not for use in diagnostic procedures.