I am using an Agilent 6460 Triple Quad and have been dealing with a terribly low signal for the past few days, the signal appears at an abundance of 30 to 50 as a baseline only. I analyzed different compounds with different concentrations, the Tune mix…
I am trying to run a Injector Precision PV test using the conditions described in the software: caffeine standard (125 ug/mL) and HP ODS Hypersil 5 um, 125 x 4 mm part 7982618-564 . The method uses 85:15 Water:ACN mobile phase at 0.8 mL/min. The test…
Hello mkaadan ,
What I meant by a "low" failure rate (notice the quotation marks), is that it is likely not an indication of a leak, as that would result in a lot more samples failing, or potentially all of them. By no means I wanted to minimize the…
Hello, H2 is the fuel for the flame in the FID, so if you are out of H2 then the FID will not be able to light. That is very likely the reason why you are not seeing a signal from the FID.
Hi ISH, yes all was fine with column installation and pump down was efficient. The issue seems to be with the entrance lense offset - I loaded a previous tune and managed to run an autotune however it is failing due to too high 503/502 ratio. I've seen…
Hello everyone,
We have a problem with the acquisition methods on GC, but not on HPLC.
When we want to change an acquisition method that is in "approved" status to "generic" status in order to modify it, it is impossible to do so because the box is…
I have decades of experience with consumer products, all on Agilent GC, GCMS, and HPLC.
If I was needing to quantitate 1,2-octanediol in cosmetics samples, I'd extract the sample with dimethylformamide and add BSTFA to make trimethylsilyl derivative…
HI Lars,
Yes, you can run the test on a VI. Specific to your system, a failing result could indicate a leak within the inlet or at any of the in-line connections of your GSV, so the results may not be clear without a point leak detector. I would also…
No, the flows on the two columns downstream of the switch have to be equal. They are feed by the same pressure source, so since MSD is under vacuum and the FID is ambient, the column connected to the MSD has to have more restriction, to keep the flows…
This Information Applies To: Agilent ICP Expert software.
Issue
The software version is often asked and is a valuable information: to see if any updates are needed, if certain options are available etc.
Steps to Follow
Launch ICP Expert...
This Information Applies To: Agilent OpenLab CDS
Issue
This article provides step-by-step instructions to calculate peak-to-valley ratio in OpenLab CDS.
Background
The peak-to-valley ratio is calculated to indicate the quality of peak separation...
This Information Applies To: Agilent 6500 Series Q-TOF LC/MS systems, MassHunter Workstation 10.1 or below and MassHunter Workstation 11.0 .
Issue
Fully reset the Q-TOF LC/MS system via a power cycle without venting. For users who experience slow...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, and Oligo Pro II systems
Issue
The volume of sample used by Agilent parallel capillary electrophoresis systems during a run.
Resolution
The...
This Information Applies To: Agilent 2100 Expert software, 2100 Expert Security Pack, 2100 Bioanalyzer
Issue
When opening the 2100 Expert software and using the Security Pack, the error message " Start of launcher application failed - The requested...
This Information Applies To: Agilent 6890, 6850, 7890, 8890, 8860, Intuvo 9000 GC systems.
Issue
The exchange of the ignitor in a Flame Photometric Detector (FPD) may be required in the following scenarios:
Preventive or periodic maintenance...
This Information Applies To: Agilent 1290 Infinity Quaternary Pump G4204A, 1290 Infinity II Flexible Pumps G7104A/C, and G7131A/C.
Issue
If the flexible/quaternary pump does not generate flow or build up pressure, it could be due to a stuck inlet...
This Information Applies To: Agilent 5200, 5300 and 5400 Fragment Analyzer systems
Issue
Separation times for the Fragment Analyzer for each array format and kit.
Resolution
Caution : The times in this article are the length of time...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, Oligo Pro II, ProteoAnalyzer systems
Issue
Guidance for installing the Agilent parallel capillary electrophoresis instrument controller and/or...
This Information Applies To: OpenLab CDS (version 2.x) system
Issue:
This article explains how to export OpenLab CDS 2.x data as Archived file.
For copying OpenLab CDS 2.x data to another Projects or another OpenLab CDS systems, you can use data...