Hi, the peak area has no influence on the calculation of Mw when measuring with concentration detectors only (RI, UV, ELSD e.g.). When performing Viscometry & Lightscattering measurements, the peak area is needed for the calculation. As an additional…
Occasionally the tune files get corrupted and you need to go back to an old one or start again from the default atune.u file. Glad the problem is resolved!
Part 1 included many questions and some answers about the maximum abundance that should be detected at the electron multiplier. Another important consideration is how much sample should go into the column.
See Which Electron Ionization Ion Source? Part…
Thank you for taking a moment to review the Community Guidelines. You agree to all of the guidelines posted and linked here when using the Agilent Community. These guidelines are here to help you be a valuable contributor to the community and make sure…
Hi, I have an old 6890 with a jet direct card. We lost power to the instrument and since then it has not been able to connect to the PC. I tried updating the bootp software with the correct mac address and power cycled instrument with the same issue.…
Support and educational resources for the Microplate Centrifuge and Automated Centrifuge Loader .
Automation Solutions Knowledge Base
Microplate-Handling Device Guides
PDF user guides
Support Documents
G5582A Centrifuge Installation Checklist…
-Have you tried a diferent column ? An old column could have retention time issues and peak distortion.. -Have you had any pressure fluctuations during your run ? Try running a Pump leak rate test, to see if the problem is not in the pump-head. -Do a…
This Information Applies To: 5200, 5300 and 5400 Fragment Analyzer, Femto Pulse and ZAG DNA analyzer
Issue
When no data is available in the Agilent ProSize data analysis software for one or more capillaries, the capillary(ies) could be misaligned...
This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
How to check the MS40+ or MS40-S foreline pump to determine if the pump needs maintenance?
Background
The oil's cleanliness and level, in the MS40+ or MS40-S foreline...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Powering on Agilent ICP-MS Systems after maintenance, long-term shutdown, or for power cycling.
Steps to Follow
Confirm that the ICP-MS power plug...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power on the Agilent ICP-MS Foreline Pump when fully powering on the Agilent ICP-MS System, or after performing maintenance.
Background
The ICP...
This Information Applies To: Agilent ICP-MS MassHunter Software version 5.1 and above
Issue
The user manuals for Agilent ICP-MS Systems for MassHunter 5.1 and above are included in the ICP-MS MassHunter Help and Learning Center. There are manuals...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Elevated oxide ratio values are defined as greater than 1.5% for x-lens configuration or greater than 3% for s-lens configuration. Possible causes are contamination...
This Information Applies To: Agilent GC/MS Systems
Issue
The software version is a valuable information usually required for troubleshooting, software updates preparation or to check installed options.
Table of contents
MassHunter GC/MS Acquisition...
This Information Applies To: Agilent LC modules with RFID tag readers
Issue
Agilent uses RFID (radio frequency identification tag) tag readers in some HPLC components such as detector lamps and flow cells, pump heads, purge valve and mixers. However...
This Information Applies To: Agilent LC systems with Lab Advisor software.
Issue
In Lab Advisor, the Logs and Results register is available for gathering history and additional information on the system. It presents data collected from the configured...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
What information does the ICP-MS instrument status indicator provide?
Background
The ICP-MS Instrument Status Indicator on the instrument, and in the...