Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-MS Portal
  • Agilent Knowledge Portal
ICP-MS Portal
Articles Elevated Oxide Ratio in ICP-MS
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-MS Portal Home
  • ICP-MS Portal Articles
  • ICP-MS Portal Videos
  • More
  • Cancel
  • Created :  3 Feb 2023
  • Modified :  4 May 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP-MS Systems
  • Component :  ICP-MS Single Quadrupole ICP MS Triple Quadrupole
  • Product Name :  7700 ICP-MS 7800 ICP-MS 7850 ICP-MS 7900 ICP-MS 8800 Triple Quadrupole ICP-MS 8900 Triple Quadrupole ICP-MS
  • Task :  Troubleshooting
Related Articles
  • Elevated Doubly Charged Ratio in ICP-MS

    This Information Applies To: Agilent ICP-MS (7700/7800/7850/7900/8800/8900) Issue Elevated doubly charged values that are greater than 3% for x-lens configuration or greater than 7% for s-lens configuration, can be caused by contamination, or an issue with the sample introduction system. Background The Doubly Charged Ratio (%) is defined as: mass 70 (Ce++) / mass 140 (Ce) x 100 Zn, Ge, ClCl, GaH, and FeO are some examples of contaminants/interferences that contribute to elevated counts...
  • How to Access Manual Tune in GC/MS MassHunter and MSD Productivity ChemStation

    Multiple Results
    This Information Applies To: MassHunter acquisition software for GC/MS systems and MSD Productivity ChemStation for all Agilent GC/MS systems Issue Manual tuning lets you interactively set the MSD parameters, such as lens voltages and tuning masses...
    • Found In:
    • Articles
    • ,
    • Articles
  • Troubleshooting an FID not Igniting

    This Information Applies To: All GCs, 6890, 6850, 7890, 7820, 8890/8860, Intuvo 9000 Issue: This article is a troubleshooting guide for when the Flame Ionization Detector (FID) does not ignite or the flame keeps going out. Cause: If the FID is having ignition issues, it is due to the following causes: FID temperature : is too low. Lit Offset : is incorrect. FID gas flows : if the air, hydrogen, or makeup gas flows are not correct (or not meeting setpoints). Gas...
  • How to Set Communication with ICP-MS in MassHunter

    This Information Applies To: Agilent ICP-MS systems (7700, 7800, 7850, 7900, 8800, 8900) Issue After shutting down and restarting the ICP-MS, sometimes the communication between ICP-MS and PC does not automatically reconnect. It may be necessary to Set Communication in MassHunter manually. Requirements The Agilent ICP-MS uses a LAN communication between the computer and instrument. Verify that the LAN crossover cable is connected to the ICP-MS and to the PC and the ICP-MS is powered...
  • Agilent ICP-MS Status Indicator

    This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900) Issue What information does the ICP-MS instrument status indicator provide? Background The ICP-MS Instrument Status Indicator on the instrument, and in the ICP-MS MassHunter Software is useful to assess the state of the instrument. It is important to monitor the status of the instrument during powering on/off, and for troubleshooting. Resolution 1. Identify the status indicator on the top-right corner...

Elevated Oxide Ratio in ICP-MS

Answer

This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)


Issue

Elevated oxide ratio values are defined as greater than 1.5% for x-lens configuration or greater than 3% for s-lens configuration. Possible causes are contamination, or a fault with the sample introduction system.


Background

The Oxide Ratio (%) is defined as: mass 156 (CeO) / 140 (Ce) x 100

Gd, Dy, CdAr, SnAr, and SeSe are some examples of contaminants/interferences that contribute to elevated counts on mass 156.

 

Items Required

  • Tuning Solution for ICP-MS 1 ppb (p/n 5185-5959), or 
  • Tuning Solution for ICP-MS 10 ppb (p/n 5184-3566)
  • Type I de-ionized water
  • Trace metal grade Nitric Acid
  • Spare Sample Introduction Components

Steps to Follow

Check for Contamination

  1. Prepare or pour fresh tuning solution in 1% HNO3 (1 ppb tuning solution for standard configuration, 10 ppb tuning solution for ISIS 3 configuration)

  2. Place the internal standard line into a fresh 1% HNO3 rinse solution (for standard configuration) or place the carrier line into a fresh 1% HNO3 rinse solution (for ISIS 3 configuration).

  3. Light plasma and perform startup tasks: Torch Axis, EM, Plasma Correction, Standard Lenses Tue, Resolution/Axis, and Performance Report.

  4. Check the oxide ratio from the startup performance report. If the oxide ratio has decreased, either the tuning solution or rinse solution was contaminated.

  5. If the oxide ratio has not decreased, it is possible that the de-ionized water source, or HNO3 source is contaminated. Prepare again the tuning solution and 1% HNO3 rinse solution with a second source of de-ionized water and HNO3, then retest.

  6. If the high oxide ratio occurred after replacing the argon, the contamination could be in the argon gas. Contact your gas supply company and test a different source of argon.

  7. It is also possible that the sample introduction system is contaminated. Clean the nebulizer, spray chamber, torch, cones, and lens stack as per the Agilent ICP-MS Hardware Maintenance Manual. See the ICP-MS MassHunter Help and Learning Center. See the article Access Agilent ICP-MS User Manuals with MassHunter 5.x for accessing the Help and Learning Center.

 

Check for Sample Introduction Fault

  1. If the high oxide ratio occurred after sample introduction maintenance, please confirm that the torch, spray chamber, nebulizer, cones, and lenses have been reinstalled properly.

  2. The oxide ratio should increase when you increase the nebulizer (carrier) gas flow. Check the nebulizer gas connection and replace the nebulizer if needed.

  3. Verify that the cones are in good condition. Poor orifice shape or size, or excessively dirty cones can cause elevated oxide ratio. Replace the skimmer and sampler cone if needed.

 
Learn how to effectively troubleshoot your Agilent ICP-MS System:
Agilent 7800/7850/7900 ICP-MS with MassHunter e-learning course available from Agilent education
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help