This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power on the Agilent ICP-MS Foreline Pump when fully powering on the Agilent ICP-MS System, or after performing maintenance.
Background
The ICP-MS MassHunter software controls the pump’s ON/OFF control, which will transition the ICP-MS System from shutdown mode to standby mode.
Steps to Follow
Figure 1. ICP-MS Rear View
1. Foreline pump breaker switch, 2. Main power breaker switch
Figure 3. Vacuum On for MassHunter 4.6 and below
Tip: If "Vacuum On" is unavailable, and there is a red X on the Hardware icon instead of a green checkmark, set communication with the ICP-MS and PC first.
|
Figure 4. Vacuum On for MassHunter 5.1 and above
Tip: If "Vacuum" is unavailable, and the vertical status bar says "Offline", set communication with the ICP-MS and PC first
|
Warning: When the plasma is lit, do not open the top covers. Confirm that the ICP-MS is in shutdown mode before continuing.
|
Figure 5. Front View of ICP-MS
1. Large Cover, 2. Manual Vacuum ON/OFF Switch, 3. LED Status Indicator
Figure 6. Top Down View of ICP-MS with Large Cover Open
1. Manual ON/OFF Switch
5. Wait for the instrument to reach standby mode. The Foreline Pump is now powered on.
Learn how to effectively troubleshoot and maintain your Agilent ICP-MS System: