I have proposed system from agilent 7890B to check high purity gases . system consists of TCD , PDHID and FID . me question is do you think that system is complicated and i have risk that if any fault happened the whole system will be down . How to guard…
Hi Kris, thank you! I tried to upload the screenshot of the version info, but I failed. So I have to type all of them out since I'm not sure which one is useful. Hope it can help!
Agilent OpenLAB Intelligent Reporting A.01.11.138
Agilent OpenLAB CDS…
Hi Carmen,
Visually check your column in signs of contamination of stationary phase. It looks like your column was too old. In that case you could just reverse the column direction, and put for for long term conditioning at room temperature, then condition…
Hello,
upon start vacuum sequence. You should hear a clik. This is for BK. Monitor IF/BK meter at that point. it has to go lower than 20 Pa. Then you should hear a second click. This time another valve open. It should rise above 20 Pa. Then after…
Hi All, I'm trying to get the following expression to work in a Qual 10 report (in excel format as that's about the extent of my coding knowledge): =IF(Area<DB_Notes,"Fail","Pass") to flag compounds which are less than a defined minimum peak area (which…
What should you consider when selecting gases for your new GC or GC/MS system?
Which gases will you be using?
Did you c onfirm that gases available are a supported configuration for installation ?
Gas requirements are specified in Site Preparation…
Hi kalle-h,
Has the method worked well for you in the past? Any changes recently like column trims? Sometimes small pieces of column can get caught inside the modulator fittings and impede flow, which has negative effects on our data.
If your peaks…
I don't think it would have a problem as long as you're not trying to connect with both software platforms at the same time. I've seen other customers do similar things (GCMS/FID, using MSD Chemstation for MSD work and GC Chemstation for FID work).
This Information Applies To: HES ion source in 5977B/C GC/MSD, 7010 triple quadrupole GC/MS systems
Issue
The filaments are replaced as part of routine maintenance, or when they break (zero output current).
Items Required
Gloves, nylon,...
This Information Applies To: Agilent 5977A/B/C GC/MSD
Issue
The Electron Multiplier Horn must be replaced when it is contaminated or the EMV value reaches 3000 V with a clean ion source.
Items Required
Electron multiplier horn (5977B/C:...
This Information Applies To: Agilent 5973/ 5975/5977 GC/MSD with rotary vane foreline pump
Issue:
The oil mist filter is recommended to be replaced annually. If the color of the filter has changed or there is smell from the filter, the oil mist...
This Information Applies To: Agilent EI Source
Issue:
During normal operation of the GC/MS, the EI SST/Inert ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filaments...
This Information Applies To: Agilent G7129A/B/C Vialsampler
Issue:
Replacing the needle and needle seat of the Vialsampler is required when:
The needle is visibly damaged.
Leaks or blockages in the vialsampler are observed, especially between...
This Information Applies To: Agilent G1311A/B/C, G7111A/B, Quaternary Pumps, and G1310A/B, G7110B Isocratic Pumps, G1312A/B/C, G7112B Binary Pumps
Issue:
How to diagnose no flow or no pressure issues on binary, quaternary, or isocratic pumps.
...
This Information Applies To: Agilent UV and RID detectors
Issue
A clogged flow cell causes high backpressure. The clog can be removed by flushing the system with solvents at high pressure. If you suspect a clog or experience high backpressure due...
This Information Applies To: Agilent G1312X and G7112X binary pumps.
Issue:
A blockage before the pressure sensor, e.g. in a pump head, will not give a high pressure or overpressure system error. Instead, the pump may show one of various error messages...
This Information Applies To: Agilent 1100, 1200, 1260 LC Pumps.
Issue:
The following procedure allows the cleaning of residue from the pump's check valves, often caused by polymerization of aged acetonitrile or salt-based solutions that may prevent...
This Information Applies To: Agilent LC Systems
Issue
After analyzing dirty or concentrated samples, some sample compounds may retain in the column and other parts of the HPLC system. If the column re-equilibration after each run is not enough to...