Problem solved. The format in Region settings should be set to "English (United States)", nothing else. As a result, the Decimal symbol under Numbers in Additional settings becomes a point (Anglo-Saxon decimalization) instead of a comma (Western European…
Hi Helene,
Thank you for contacting Cell Analysis Technical Support. Sodium bicarbonate cannot be used because it is a buffering agent with a very high pKa and, unless incubated in a CO2 incubator, will raise the pH of the assay medium to well above…
Great. The 219 and 502 are quite high. Make sure that the extractor lens insulator is clean, the extractor pin isn't touching the body, the brown extractor lens connector wire is connected well and not crimped, and that the transferline tip seal ceramic…
Hi AVEdsell,
Site prep clearly states that system MUST be vented, this is a snippet from the TQ site Prep Guide about Venting. I do not know waht MS you have but this is common for all :
Bonjour Fabrice,
Juste pour ajouter a l'information que Kristen a deja donne:
Si vous selectionne la langue comme Francais, vous pouvez trouver le manuel. La description, de mon cote, a l'air d'etre en anglais, mais si vous cliquer sur le lien…
Hello lperry ,
For Cool on Column inlets, the septum and the inserts types are dependent on the column size. It is quite possible that either the septum, the insert, or both, are not matching the column size.
Are the needles getting bent too? Does…
Hello Samehalsharawy ,
Can you please elaborate on the statement we uninstalled the cylinder expecting the flow to come down to zero but it didn't. It stayed at 50?
If no physical tank is connected to the GC, the GC gas connections are capped…
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue
The Agilent 2100 Bioanalyzer system run ends prematurely and the following error appears in the logbook: Run aborted on port x. Instrument error occurred on port x, Unusual...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue
The Agilent 2100 Bioanalyzer system electropherograms in the Agilent 2100 Expert software have a wavy, uneven baseline ( Figure 1 ).
Figure 1. Wavy, uneven baseline on Bioanalyzer...
This Information Applies To: Agilent AriaMx systems.
Issue
When IP addresses of an AriaMx instrument are reassigned, it will appear under the new and the previous IP address in the AriaMx Instrument Explorer. This situation can be prevented using...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Sometimes, for support purposes, it is necessary to send the diagnostic test result file to Agilent support representatives.
This file contains detailed information about the...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx Real-Time PCR software
Issue
How to change the discovery protocol port in the AriaMx software.
Steps to Follow
The AriaMx software uses the UDP Broadcast protocol for...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
The instrument diagnostic tests should be run whenever it is suspected that there could be a fault with the instrument hardware.
Background
The instrument diagnostics consist...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Error 45 is displayed on the AriaMx instrument touch screen.
Background
Error 45 is shown when the temperature recorded by one or more thermistors is outside the expected...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx showing a series of errors after delivery, guide for troubleshooting and resolving temperature-related error.
Background
When an AriaMx is delivered, it could take...
How to Access Manual Tune in GC/MS MassHunter and MSD Productivity ChemStation
Multiple Results
This Information Applies To: MassHunter acquisition software for GC/MS systems and MSD Productivity ChemStation for all Agilent GC/MS systems
Issue
Manual tuning lets you interactively set the MSD parameters, such as lens voltages and tuning masses...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
How to find the AriaMx software and firmware version that you have installed.
Steps to Follow
There are two different ways to find out AriaMx software version:
In the...