How do you split the sample? Is it two different injection ports? Two different columns into the injection port? Or a two way splitter - purged or unpurged?
Are you injecting split or splitless? Please upload the acqmeth.txt file that is in Windows…
Hello,
Error 1213 typically comes with other errors that truly describe the issue, like the 2150 error you are getting.
First, check the cone orifices. Use a microscope or powerful magnifying glass. If they are clogged or salted over, clean. The opening…
This may be due to several factors. First of all, make sure that all plasma monitor connections and cables are properly connected and not damaged. Also check that there is sufficient gas flow (eg argon) to maintain the plasma. Check the cleanliness and…
Hello all,
I´m facing a Silicium contamination in ICP-MS analysis, I have used a 7800 ICP-MS.
I have digested my samples with HF, HCL and HNO3 and neutralized it with boric acid, after the analysis I have Silicium CPS has increased around 30%.
…
Hi Layden,
your source has died (normal ratio Amps/Volts should be starting at 1.9/2.7). Each year you will find a decline in Amps until it has reached 1.6, then it will give a red bar. In your case it's 0, so the filament is probably broken - your…
I have almost the same system as shown in the picture for over a year now. The application was supplied by Da Vinci Laboratory Solutions in the Netherlands. It is a Mettler TGA 2, with a Gerstel MPS autosampler on a Agilent GCMS(MS) with a CIS injector…
Thank you. I contacted the support. I just had to login into agilent subscribenet and download the latest version after entering the authorization key.
Hello,
That looks like a standard IS calibration using area percent. The purity of the standard can be entered as the compound multiplier. I would start by searching the help for calibration, internal standard, and mass percent calculations. You can…
This Information Applies To: Agilent AriaMx, Software version 1.8 and later
Issue
The Aria Warm up Feature was introduced with Aria SW/FW 1.8 to reduce baseline drift. The warm-up feature is used to speed up the heating process that would normally...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
When a sample name starts with zero, the leading zero disappears from the exported data ( Figure 1 and Figure 2 ).
Background
Removing the leading...
This Information Applies To: Agilent 7890B GC system and 7697A headspace sampler
Issue
The 7697 Headspace Sampler can be configured to communicate directly with the 7890B GC. When this direct communication is configured, the 7697A headspace gets...
This Information Applies To: Agilent GC/MSD systems
Issue:
When performing MS tuning, there is error message "No Emission Current" and tune is aborted.
Steps to Follow
This error normally means that the selected filament is not working. Perform...
This Information Applies To: Agilent OpenLab CDS
Issue
Any desired processing method can be assigned to the current Result Set.
Requirements
A result set available to link the method to.
Correct Privilege (Role Membership) set up for the...
This Information Applies To: Agilent LC and CE Systems compatible with Lab Advisor
Issue:
Firmware is updated or downgraded when new features are added, or to fix some system bugs. It also solves compatibility issues when a new module is added to...
This Information Applies To: Agilent OpenLab CDS
Issue
This article shows how to export a report of peak area and retention time of a sequence in an Excel format.
Prerequisites
To be able to carry out the procedure as described, privileges...
This Information Applies To: Agilent OpenLab CDS
Issue
Reports in OpenLab CDS can be printed or saved manually in Data Analysys, or automatically during reprocessing, or acquisition.
Requirements
The user needs the following Roles and Privileges...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD systems.
Issue
During tuning, MassHunter prompted the error message: "MSD Fault- RFPA Difficulty", and tune action stopped.
Steps to Follow
Try to reconfigure the GC/MS....
This Information Applies To: Agilent OpenLab CSD 2.5 and above
Issue
The integration optimizer helps find a suitable set of integration events. It allows integration parameters to be tuned via sliders (peak width, slope sensitivity, height reject...