Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Chromatography Software Portal
  • Agilent Knowledge Portal
Chromatography Software Portal
Articles How to Print Reports in OpenLab CDS
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Chromatography Software Portal Home
  • Chromatography Software Portal Articles
  • Chromatography Software Portal Videos
  • More
  • Cancel
  • Created :  21 Apr 2023
  • Modified :  1 Sep 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical Software Suite
  • Component :  OpenLab Chromatography Data System
  • Product Name :  OpenLab CDS
  • Task :  Operation Reporting
Related Articles
  • How to Add and Set Up a Report Line to the Sequence Table in OpenLab CDS

    This Information Applies To: Agilent OpenLab CDS Issue This article will explain how to add a sequence summary report line to a sequence, in Agilent OpenLab CDS, as well as how to configure it so it matches your reporting needs. Background Adding a sequence summary report line to the sequence requires a basic understanding of sequences as well as how to submit runs and sequences within OpenLab CDS. For further detail on this subject, read the following article: How to create, edit and...
  • Filtering Sample Type Using Enumeration in Reporting with OpenLab CDS

    This Information Applies To: Agilent OpenLab CDS Issue This article outlines step-by-step instruction to filter sample type using Enumeration in Reporting. Prerequisites The privilege Data Processing > Print results reports is required. This privilege is configured in the Control Panel . Sample type must be defined in either Sequence Table in Acquisition ( Figure 1 ) or Injection List in Data Analysis ( Figure 2 ). Figure 1 . Sample Type Column in Sequence Table Figure 2 . Sample...
  • Reporting Peak Area and Retention Time of Sequence in Excel Format with OpenLab CDS

    This Information Applies To: Agilent OpenLab CDS Issue This article shows how to export a report of peak area and retention time of a sequence in an Excel format. Prerequisites To be able to carry out the procedure as described, privileges of Data Processing > Print results reports and Report Template > Create report template are required. Privileges are configured in the Control Panel . The relevant peaks in the chromatograms are integrated. Steps to Follow Identify peaks...
  • Generating a Report in MassHunter Qualitative Analysis Software

    This Information Applies To: Agilent MassHunter Qualitative Analysis Software version 10.0 Issue The Agilent MassHunter Qualitative software can generate qualitative analysis reports, which may include sample chromatogram information and sample spectrum information, library search details, system suitability parameters, etc. The following steps demonstrate how to generate a custom report in the Navigator view. Steps to follow 1. The procedure starts with loading the required data, selecting...
  • How to Review and Confirm Audit Trails in Data Analysis by Administrators at OpenLab CDS (version 2.x)

    This Information Applies To: OpenLab CDS (version 2.x) Systems Issue: This article explains tips for how to review and confirm audit trails of Result Set / Injection Data / Processing Method by Administrators at Data Analysis of OpenLab CDS (version 2.x). Tips : Reviewing the audit trails requires the role privilege Review Audit Trail . The Administrative User can set the role privilege from the OpenLab Control Panel. If this privilege is not predefined, the Review button cannot...

How to Print Reports in OpenLab CDS

Answer

This Information Applies To: Agilent OpenLab CDS 


Issue

Reports in OpenLab CDS can be printed or saved manually in Data Analysys, or automatically during reprocessing, or acquisition.


Requirements

The user needs the following Roles and Privileges for printing reports: Data > Save reports to disk and Data Processing > Print results reports.

For detailed steps on where to set up the Roles and Privileges, see the article: How to Create or Edit Roles in OpenLab Chromatography Data System.

If you want to be able to print unsaved results, you need to check the box indicated in the Figure 1, within the Projects section of Control Panel.


Enabling print unsaved results.
Figure 1. Enabling print unsaved results.

Steps to Follow

There are two ways to print reports from Data Analysis:

  1. Using the Processing Method parameters in Data Processing tab
  2. Using the Print to Printer icon in Reporting tab

Using the Processing Method parameters in Data Processing tab

This option is used for individual runs or method reports.

  1. In the Home tab, click the Method layout (Figure 2).


    Location of the Method menu.
    Figure 2. Location of the Method menu

  2. Click Injection Report parameter, under the Reports drop-down (Figure 3).


    KP 937 Figure 2
    Figure 3. Location of the Injection Report field, within the Processing Method.

  3. Complete the fields as per your analytical needs. See Figure 4 for reference.

      • Report Template: Browse the report template to use for this result set.
      • Report Selected Signals: Select whether you want to report only some signals, or all of them.
      • Report for Sample Types: Select to report only a particular sample type (for example, just the calibration samples) or all of them.
      • Report Destination: Select how you want to save the report as a file, and/or send the report to a local printer.
      • Report File Name: Choose a name format for the report file that you want to save/print.
      • File Format: Select the file formats in which you want to save your file under (if applicable)
      • Copy Report to Folder: Select whether you want to save a report to a folder or not.
        • If you select Storage, the file is saved under the current project.
        • If you select Windows File System, the file can be saved anywhere on your PC.
      • Destination Folder: Displays the location where the file is saved, based on the options chosen in the previous step.


        Fields available for defining the Report parameters.
        Figure 4. Fields available for defining the Report parameters

  4. in the top ribbon, click Reprocess All (see Figure 5).

  5.  

    Click Save All Results. 
    If you do not save the results after reprocessing, you will get the following error when trying to print (Figure 6).


    Location of the Reprocess All and Save All Results menu items.
    Figure 5. Location of the Reprocess All and Save All Results menu items


    Warning displayed when printing without saving results.
    Figure 6. Warning displayed when printing without saving results

  6. Click the Print All icon (Figure 7).


    KP 937 Figure 6
    Figure 7. Location of the Print menu icon


      
    Confirmation: The results are now printed at the desired printer, if applicable, and the selected file types are saved at the location indicated under the Destination Folder option.

Using the Print to Printer icon in Reporting tab

  1. Load the desired report template from the Report Template drop-down (Figure 8).


    Report Templates available for use.
    Figure 8. Report Templates available for use

  2. Click Preview. Steps 2 and 3 are shown in Figure 9

  3. Click the print icon under the Report Preview window.


    KP 937 Figure 8
    Figure 9. Location of Print Preview and Print menu options.

      
    Tips: If you made any changes to the report, click Refresh Preview in the top ribbon to see the recent changes.

      
    Confirmation: You will now be directed to selecting a printer; if applicable, you will be also directed to choose a destination folder for saving the results as a PDF.

There is also another way to print reports, directly from the Sequence Table during method acquisition. This option assumes that the processing method does not need further modifications and it can be used as is. However, if you need to adjust integration parameters, integration events, or do any sort of addition/modification to the processing method, you need to use the two steps outlined in this article.

  Learn how to effectively operate your Agilent OpenLab CDS:
Agilent 8890 GC with OpenLab CDS
Agilent 1260 Infinity LC Systems with OpenLab CDS e-learning paths available from Agilent education