This Information Applies To: Agilent 7890B GC system and 7697A headspace sampler
Issue
The 7697 Headspace Sampler can be configured to communicate directly with the 7890B GC. When this direct communication is configured, the 7697A headspace gets the 7890B GC column configuration, oven program, synchronizes with the GC clock, and follows the GC instrument schedule, such as wake-up and sleep.
Requirements
This feature requires a headspace sampler with firmware A.01.06 or greater, and a 7890B GC with firmware B.02.02 or greater. If necessary, see instructions on updating Agilent GC Firmware.
Steps to Follow
2. At the GC keyboard
If the communications fail, check the IP addresses
Learn how to effectively operate your Agilent 7697A Headspace: