Hi ISH, yes all was fine with column installation and pump down was efficient. The issue seems to be with the entrance lense offset - I loaded a previous tune and managed to run an autotune however it is failing due to too high 503/502 ratio. I've seen…
HI Lars,
Yes, you can run the test on a VI. Specific to your system, a failing result could indicate a leak within the inlet or at any of the in-line connections of your GSV, so the results may not be clear without a point leak detector. I would also…
Hello Paul, it looks to me that there is too much acetylene in the gas mixture coming to the burner. This is causing the carbon build-up.
I'd suggest you first remove the absorption cell from the holder and clean this by soaking in dilute nitric acid…
Hi Mohammad, Thanks for your thought, detector temperature is not extremely high or low. I did monitor temperature for FID and inlet the whole time even during the injection. It was stable and maintained for hours/days. Method that I was using I had checked…
I’m sorry for the delayed reply to your message. Do you still need assistance with this? If you do, please take into consideration the notes below. However, if the issue was resolved, could you please share what the resolution was?
From what I understand…
Hi all,
I have a 6890A GC with a FID. We recently acquired the instrument and I have been running standards up to test methods. I have been getting signals that seem suspiciously low.
I am injecting Matreya Bacterial CP FAME mix (C11-C20 methyl esters…
Hello everyone,
I have a 8890 GC-FID that I've just started running.
I want to create a "shut down" method that turns off (or lowers) H2 and Air once a run is completed or that can uploaded at the end of the day. The goal here is to keep the instrument…
I have a 6490 QQQ. Each time I run either autotune or check tune, I get masses that are out of tolerance especially in MS2. The system seems to pass if the Dynamic Ramp Table enable button is checked. This is associated with the collision cell and detector…
Based on your latest results, I would say that nothing is fundamentally wrong with the optical unit in terms of noise, so I believe this is more an applicational problem - your passed quick noise test points to that direction. The error 1050 with the…
Disconnect the sample probe from the valve and start the leak and restriction test. If it passes this step, it means that there is some restriction in the sample probe, enough to cause that part of the test to fail. Try a brand new probe. If it fails…
This Information Applies To: Agilent MassHunter Qualitative Analysis Software
Issue
Error message prompted “Access to the path C:\Users\AgilentMad\AppData\Local\Temp\FILESPEC.TXT is denied" during search using NIST MS Program in MassHunter Qualitative...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
When sending a run configuration from the AriaMx PC software, no run configuration is transmitted. The expected blue confirmation icon does not appear at the bottom of the touch...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
Can the AriaMx software be used on Windows 11?
Resolution
The AriaMx a nalysis software 2.0 has not been validated for Windows 11, so Agilent...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
How to control more than one AriaMx with one PC.
Resolution
If you connect the PC and instruments using an ethernet cable, you will not be able to establish multiple connections...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
In the Export Data menu, after exporting data in a LIMS format, the output file's name is incomplete.
Background
An example of everything after...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
The AriaMx Data saving behavior when priming experiments from the PC is dependent upon whether the user is logged in or a Guest and whether the user...
This Information Applies To: Agilent OpenLab CDS
Issue
Sequence Creation Template allows users to create a sequence template that identifies the order of injections or actions in the sequence.
This article outlines the step-by-step instruction...
This Information Applies To: Agilent 6100 and 6400 Series LC/MS Systems
Issue
Tuning a mass spectrometer involves optimizing the voltages applied to various optical elements in the ion path to achieve the required mass accuracy, resolution, and...
This Information Applies To: Agilent OpenLab ChemStation
Issue
In OpenLab ChemStation, creating a result set by combining multiple existing result sets can be useful in the following scenarios:
Combining single samples, sequences, or a combination...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
The AriaMx analysis software shows an error when attempting to open a data file.
Background
The AriaMx analysis software can show several errors when something is wrong...