Get a WS asynchronous error : F F60304 The torch ignition has failed when starting a Varian 720 OES The torch lights and stays on for about 5 seconds then shuts down with this error. I can not find reference in the help to resolve this. Thanks, Scott
Using the Agilent 7890A. Column installed with no leaks detected. Both the front and back detector are set to 300°C, gas Flow Rates - Hydrogen: 45mL/min, Air 450mL/min and He 1.0mL/min. The split flow is 80mL/min. Following initial start up the front…
Hi John, You are on the right track-- it sounds like your flame is not lit. The reading you are seeing on the Hydrogen line is likely back-pressure from the column flow. That result could indicate a blockage in the jet, but the flow is a little low to…
Thanks for the tips on seeing the position in the sequence once Aquisition has been restarted. Unfortunately something I failed to mention is that the sequence also disappears after pressing run but will re-appear once you toggle between another module…
I have started using GC-FID (Agilent 8890) since March for the first time. We are trying to measure the methane. I was trying to make a standard surve, so I injected 1000 ppm methane standard at first. Just to let you guys know no samples has been run…
It looks like it is having to clear something out the inlet, column, detector or head space as the baseline rise is over most of it. I see it's a FID you are using, how long do you leave it ignited before running the first analysis?
no you don't need to disconnect it from the FID, ,
bake the column for enough time usually more than 1 hour according to your column status , you can keep the FID working to monitor the baseline .
..
keep temperature 25 degrees less than the maximum…
External Device not ready. Is it an Agilent Autosampler? Which model? Do you have another device like a headspace sampler?
APG Remote. There is a cable between the GC and the MSD so that they both start at the same time. A remote start cable from another…
I am running on an Infinity 1260 II Ultivo LC/QQQ and have a problem when running check tunes. Every time I run a check tune, I get an out of tolerance result where the only parameter out of range is the gain factor. It fluctuates between being out of…
Hi, I'm working with a LC Infinity Lab 1260, which has a standard autosampler and 100 uL loop. I'm trying to achieve a very simple injection program, but keep failing. All I want to do is to draw 50 uL from a sample vial, 50 uL from a second sample vial…
This Information Applies To: Agilent MassHunter Qualitative Analysis Software
Issue
Error message prompted “Access to the path C:\Users\AgilentMad\AppData\Local\Temp\FILESPEC.TXT is denied" during search using NIST MS Program in MassHunter Qualitative...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
When sending a run configuration from the AriaMx PC software, no run configuration is transmitted. The expected blue confirmation icon does not appear at the bottom of the touch...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
Can the AriaMx software be used on Windows 11?
Resolution
The AriaMx a nalysis software 2.0 has not been validated for Windows 11, so Agilent...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
How to control more than one AriaMx with one PC.
Resolution
If you connect the PC and instruments using an ethernet cable, you will not be able to establish multiple connections...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
In the Export Data menu, after exporting data in a LIMS format, the output file's name is incomplete.
Background
An example of everything after...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
The AriaMx Data saving behavior when priming experiments from the PC is dependent upon whether the user is logged in or a Guest and whether the user...
This Information Applies To: Agilent OpenLab CDS
Issue
Sequence Creation Template allows users to create a sequence template that identifies the order of injections or actions in the sequence.
This article outlines the step-by-step instruction...
This Information Applies To: Agilent 6100 and 6400 Series LC/MS Systems
Issue
Tuning a mass spectrometer involves optimizing the voltages applied to various optical elements in the ion path to achieve the required mass accuracy, resolution, and...
This Information Applies To: Agilent OpenLab ChemStation
Issue
In OpenLab ChemStation, creating a result set by combining multiple existing result sets can be useful in the following scenarios:
Combining single samples, sequences, or a combination...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
The AriaMx analysis software shows an error when attempting to open a data file.
Background
The AriaMx analysis software can show several errors when something is wrong...