I have tried two new lamp and its all fail intensity test on the controller. Then I borrow a standard flow cell from another LC and I got min. intensity pass but fail average intensity. So I use Methanol to clean the flow cell window and flush the flow…
Hello, This issue sounds like the one described in KPR#:245690 of the OpenLab ChemStation Software Status Bulletin: https://www.agilent.com/cs/library/support/Patches/SSBs/M83xxAA.html#TTID:245690 I have copied the contents of the relevant entry below…
Hello, I have a Agilient CP7420 column connected to the GCMS-QP2020NX also equipped with a FID detector and the column is connected to the FID (carrier gas helium 1mm/min). The column and the chromatograph are new, but the column was purchased from 3…
Hello Nbauer, thank you! I've tried to move the blanking nut at the outlet valve but the result is the same "test aborted by system" and, on the right "failed to comunicate with the module". All modules are connected so maybe could be a problem with the…
Hi Paul,
Thanks for the response. We are using Agilent 5183-4702 liner, HP-5ms Ultra Inert Columns (Part Number:19091S-433UI, 30 m, 0.25 mm, 0.25 µm, 7 inch cage).
We are running in Scan, absolute. I always use autotune, I didn't find a place to choose…
Dear Sir/Madam
Good morning. Currently, I am working on GC-MS, Agilent. I am able to run my samples as a batch samples. However, I am not able to quantify as well as prepare calibration curve using MassHunter. I followed the user manual but failed.…
Hi!
I run 200.8 on the 7850 ICPMS for drinking water.
I calibrate with 10, 50, 100, 250, and 500 ppb for Cu (among others). I would like to calibrate up to 2,000 or higher for Cu - but every time I try to calibrate above 500, my CCV and LFB fail…
Hi, We have agilent 6545 qtofms. The quad tune in neg mode is failing due to failed wide peak width for 1633 ion. any suggestions to troubleshoot. please see the image. Thanks
Hello
We have an agilent ultivo qqq and last week when trying to run a checktune it failed giving a message saying calibrant bottle is empty or nebulizer is clogged. Nevertheless the calibrant bottle is definitely not empty, and the nebulizer also doesnt…
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...