I'm not aware of any way to do this with just one compound/cal curve. One solution would be to make duplicate compounds and assign one compound the low level standards and the other the high level standards. You can utilize different…
Thank you, unfortunately we tried to set a collection window much larger than the peak duration, but the peak got split again. And in the version of the software we are using no advanced settings allowing to collect different peaks into the same vial…
Hi CMZainea,
Thank you for your feedback. The leak test for the 6 ports valve was passing, so no leaks detected at the valve. Any other leaks were detected from the visible piping and connection from the back panel of the 7697A going to the 6 ways valve…
The two-way splitter without makeup G3181B can be used as a splitter or a combiner to connect two inlets to one detector. Discuss this part with your local sales rep and they can help you get a quote for this part.
https://www.agilent.com/cs/library…
The Digest of upcoming webinars has been updated as of February 20.
Notable additions include:
BioPharma (1)
GC & GCMS (2)
Pharma (1)
Software & Informatics (21)
Check Out:
Tips and Tricks Workshop on ICP-OES: Smarter Method Development…
This Information Applies To: Agilent 5100, 5110, 5800, and 5900 ICP-OES systems
Issue
When troubleshooting communication issues between Agilent 5000 Series ICP-OES and the instrument controlling computer, it is necessary to know the IP address of...
This Information Applies To: Agilent 5000 series ICP-OES systems
Issue
Poor or high relative standard deviation, or a noisy signal is an indication of poor precision during analysis.
Resolution
There are many potential causes for an instrument...
This Information Applies To: Agilent 5000 series ICP-OES systems
Issue
Instrument calibration (wavelength), detector calibration (dark current, or method calibration (worksheet) failure.
Resolution
There are several reasons why a calibration...
This Information Applies To: Agilent 5000 series ICP-OES
Issue
Exhaust temperature and requirements for laboratory site preparation
Background
Certain parameters will impact the actual temperature of the plasma exhaust. Some of these factors...
This Information Applies To: Agilent 700 Series ICP-OES instruments
Issue:
It is important that a torch alignment is performed to ensure that the optics are optimized and are focused on the analytical zone of the plasma. This alignment procedure...
This Information Applies To: Agilent 5800/5900 ICP-OES
Issue:
The instrument requires clean, dry, noncorrosive air for cooling purposes. The air enters the instrument through an air supply vent located beside the chimney on the top of the instrument...
This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES systems
Issue
The ICP-OES spray chamber filters the sample droplets created by the nebulizer. Larger droplets fall to the chamber drain, while the smaller droplets are introduced...
This Information Applies To: Agilent 5000 ICP-OES Series and Agilent ICP Expert Software
Issue
How to connect Agilent 5000 series ICP-OES systems to Agilent ICP Expert software via LAN
Steps to Follow
How to connect Agilent 5000 series ICP...
This Information Applies To: 5000 series ICP-OES (5100/5110/5800/5900), ICP Expert
Issue
How to Reset the AVS 6/7 after maintenance or when a red LED is visible
Steps to follow
Tips : Instructions for changing the AVS switching...
This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES
Issue
The axial pre-optic window needs checking from time to time to ensure it is clean.
If the axial pre-optic window is dirty, it must be cleaned or replaced to ensure adequate...