Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Cleaning the Agilent ICP-OES Spray Chamber
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP OES Systems
  • Product Name :  5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES
  • Task :  Maintenance
Related Articles
  • Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES

    This Information Applies To: 5100, 5110, 5800, and 5900 ICP-OES instruments Issue The ICP-OES operator can identify the source of plasma ignition failure. The main cause is air entering the plasma. Steps to follow Excessive air entering the plasma. Check the following: The peristaltic pump tubing is clamped and has sufficient clamping pressure. Usually, air is being pulled into the nebulizer because the pump tubing is not properly clamped during ignition. At ignition, the plasma...
  • Cleaning the ICP-OES Nebulizer

    This Information Applies To: Agilent 500 Series ICP-OES (5100, 5110, 5800, 5900) Issue The ICP-OES nebulizer converts the liquid sample into an aerosol as it enters the spray chamber. A dirty or partially blocked nebulizer traps sample, which leads to poor reproducibility and reduced sensitivity. Droplets on the internal surface of the spray chamber can indicate poor conversion of liquid to mist. ICP-OES overview The ICP-OES sample introduction components are detailed in Figure 1 . ...
  • Troubleshooting a Leak Error in a Fluorescence Detector

    This Information Applies To: Agilent Fluorescence Detectors G1321A, G1321B, G7121A, G7121B Issue This article describes how to diagnose a leak in your fluorescence detector (FLD) module. Steps to Follow Confirm the leak error originated in the FLD module before starting this procedure. See Determining which Module is Causing the Shutdown Error Due to a Leak Warning : The following procedure risks exposure to solvents in the flow path of the FLD flow cell. Always wear safety...
  • Cleaning the 5000 Series ICP-OES Torch, Acid and Aqueous Matrices

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) Issue The ICP-OES torch plasma ionizes the introduced gas or aerosol stream. In normal operating conditions, incomplete ionization occurs and material is deposited on the torch surface. Build up of deposits can create blockages, damage the torch or interact with the sample stream causing poor precision, accuracy, and memory effects. ICP-OES overview The ICP-OES sample introduction components are detailed in...
  • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) and Agilent 700 Series ICP-OES (710/715/720/725/730/735) Issue Loss or change of sensitivity in ICP-OES can occur under normal operating practices. This discussion identifies factors that can contribute to sensitivity changes including calibration standards , memory effects , and the sample introduction system . Sample introduction components include: Autosampler probe Switching valve Pump tubing Nebulizer...

Cleaning the Agilent ICP-OES Spray Chamber

Answer

This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES systems


Issue

The ICP-OES spray chamber filters the sample droplets created by the nebulizer. Larger droplets fall to the chamber drain, while the smaller droplets are introduced to the gas flow going into the ICP torch plasma. A dirty spray chamber causes large droplets to form, which leads to incomplete transfer of sample to the torch, causing poor precision and memory effects. 


ICP-OES overview 

The ICP-OES sample introduction components are detailed in Figure 1.

   
Figure 1: ICP-OES sample introduction components
1. Torch loader handle 2. Torch 3.Torch clamp 4. Nebulizer gas connections 5. Spray chamber drain tubing 6. Spray chamber 7. Nebulizer 8. Nebulizer inlet tubing

How to identify a dirty spray chamber

 Tip Routine analysis of aqueous and nonaqueous samples using a single spray chamber will increase cleaning and instrument downtime. Agilent recommends the use of two spray chambers, one for aqueous solutions and one for organic solutions, to reduce downtime. 

A clean spray chamber has a uniform mist on the internal surface (Figure 2). A dirty spray chamber has large droplets on the internal surfaces (Figure 3).


Figure 2: Correctly wetting spray chamber

 
Figure 3: Incorrectly wetting spray chamber

Steps to follow

  1. Turn off the pump and plasma.
     Warning
    Warning: The spray chamber may contain potentially dangerous chemicals, and the nitric acid and alkali detergent wash solutions are corrosive. Always wear Personal Protective Equipment when cleaning the spray chamber.
  2. Remove the nebulizer (see Figure 1).
  3. Remove the spray chamber and disconnect the tubing. This video demonstrates how to remove the spray chamber:
  4. Rinse the spray chamber with 1% HNO3 solution then ASTM Type 1 water. 
  5. Prepare an aqueous solution of a quality laboratory alkaline detergent. Completely immerse the spray chamber in the solution, ensuring all the internal surfaces of the spray chamber are in contact with the detergent solution.
  6. For:
    • Single pass spray chambers: The container may be placed in an ultrasonic bath and sonicated for 40 minutes.
    • Multiple pass spray chambers: Do not sonicate, instead allow the spray chamber to stand overnight in the detergent solution.
       Caution 
      Caution: Sonicating a multiple pass spray chamber may cause damage.
  7. Remove the spray chamber from the detergent solution and rinse thoroughly with ASTM Type 1 water.
  8. Dry external surfaces using a clean, lint free paper towel. Do not dry the spray chamber using compressed gases.
  9. Re-attach all tubes, adaptors, and the nebulizer. Reconnect the assembly to the instrument.
  10. Test the cleaned spray chamber by nebulizing a blank solution and repeat the cleaning process if there are large droplets on the spray chamber internal surfaces.
     Tip 
    Tips: It is not unusual that some spray chambers will require extended cleaning.
  11. If the spray chamber is not going to be used immediately after cleaning:
    • Store in an appropriate covered vessel containing a 1% HNO3 solution until required or,
    • Store the dried spray chamber in its original packaging.
Tip Recommended maintenance schedule:  Agilent recommends weekly to monthly cleaning of the spray chamber. Analysis of samples with complex matrices may require more frequent cleaning. 

 Tip 

Learn how to effectively maintain your Agilent ICP-OES:
ICPOES-5110-1270e - Agilent 5110 ICP-OES Maintenance and Troubleshooting  e-learning course available from Agilent education

Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Privacy Statement
Terms of Use
Contact Us
Site Help