This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES systems
Issue
The ICP-OES spray chamber filters the sample droplets created by the nebulizer. Larger droplets fall to the chamber drain, while the smaller droplets are introduced to the gas flow going into the ICP torch plasma. A dirty spray chamber causes large droplets to form, which leads to incomplete transfer of sample to the torch, causing poor precision and memory effects.
ICP-OES overview
The ICP-OES sample introduction components are detailed in Figure 1.
How to identify a dirty spray chamber
Routine analysis of aqueous and nonaqueous samples using a single spray chamber will increase cleaning and instrument downtime. Agilent recommends the use of two spray chambers, one for aqueous solutions and one for organic solutions, to reduce downtime. |
A clean spray chamber has a uniform mist on the internal surface (Figure 2). A dirty spray chamber has large droplets on the internal surfaces (Figure 3).
Figure 3: Incorrectly wetting spray chamber |
Steps to follow
Warning: The spray chamber may contain potentially dangerous chemicals, and the nitric acid and alkali detergent wash solutions are corrosive. Always wear Personal Protective Equipment when cleaning the spray chamber.
|
Caution: Sonicating a multiple pass spray chamber may cause damage.
|
Tips: It is not unusual that some spray chambers will require extended cleaning.
|
Recommended maintenance schedule: Agilent recommends weekly to monthly cleaning of the spray chamber. Analysis of samples with complex matrices may require more frequent cleaning. |
|
Learn how to effectively maintain your Agilent ICP-OES:
ICPOES-5110-1270e - Agilent 5110 ICP-OES Maintenance and Troubleshooting e-learning course available from Agilent education
|