Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles How to Set up Communication Between Agilent 5000 Series ICP-OES Instruments and Agilent ICP Expert Software
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP-OES Software ICP OES Systems
  • Product Name :  5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES ICP Expert software
  • Task :  Operation
Related Articles
  • How to Power Cycle Agilent 5000 ICP-OES Systems

    This Information Applies To: The Agilent 5000 ICP-OES Series and Agilent ICP Expert Software Issue How to reset the Agilent ICP-OES system when experiencing a communication error between the instrument and software, or if the instrument seems to be experiencing issues when trying to light the plasma or run a worksheet. These steps will help to determine if there is a hardware issue with the instrument. Steps to Follow Caution : Save all data in ICP Expert software before shutting...
  • How to Upgrade ICP Expert version 7.5

    This Information Applies To: 5000 series ICP-OES (5100/5110/5800/5900) ICP Expert version 7.5 Issue How to install an upgrade version of ICP Expert via USB to current ICP Expert (i.e. 7.5.2 upgraded to 7.5.3) Resolution To upgrade the current version of software to the new version of software, please have your current ICP Expert Software closed and log onto the computer with Administrator privileges. 1. Insert the ICP Expert software USB in the PC 2. If AutoPlay prompts to Run ICPExpertSetup...
  • AVS 6/7 Tubing Connections and Part Numbers

    This Information Applies To: Agilent 5000 Series ICP-OES, AVS 6/7 Issue This article shows the required t ubing connections and part numbers between the AVS 6/7 and ICP-OES ICP-OES sample introduction system: Figure 1. Sample introduction system and AVS 7 (shown as 1) Figure 2. Connections to AVS: 1 to AVS Pump, 2 from Autosampler probe (rinse), 3 Argon gas, 4 Rinse/Carrier Figure 3. AVS (showing tubing removed): (A) AVS valve, (B) Sample loop holder, (C) AVS pump inlet, (D) AVS...
  • Replacing the AVS 6/7 Rotor

    This Information Applies To: Agilent 5000 Series ICP-OES Systems, AVS 6/7 Issue Over time the AVS 6/7 rotor can become worn or damaged by chemicals. Agilent recommends inspecting and/or replacing the rotor every 5000 AVS switches or when analytical results or precision fall outside expected limits. Steps to Follow: Tip : Refer to the Help section of the Agilent ICP Expert Instrument Software for further detailed instructions. Warning : Always wear appropriate...
  • SPS 4 Alignment and Calibration Wizard

    Multiple Results
    This Information Applies To: Agilent Spectroscopy Systems, SPS 4 Autosampler Issue The sample probe may be displaced from its correct position due to the probe hitting a sample vial or the sides of the wash reservoir. User calibration allows for...
    • Found In:
    • Articles
    • ,
    • Articles
    • ,
    • Articles

How to Set up Communication Between Agilent 5000 Series ICP-OES Instruments and Agilent ICP Expert Software

Answer

This Information Applies To: Agilent 5000 ICP-OES Series and Agilent ICP Expert Software


Issue

How to connect Agilent 5000 series ICP-OES systems to Agilent ICP Expert software via LAN


Steps to Follow

How to connect Agilent 5000 series ICP-OES instruments to Agilent ICP-Expert software using the standard configuration of a static IP address

The Agilent 5000 series ICP-OES uses a LAN communication between the computer and instrument. 

  Caution
Caution: Any changes to default network settings will not be accessible by Agilent Representatives. This procedure details the standard configuration using a static default IP address for the instrument, and does not take into account any additional network settings or connections.
  1. Connect the LAN cable between the computer and instrument.  
    1. The ICP-OES connection port is on the right side of the instrument
  2. Set up the static IP address on the computer
    1. On Microsoft Windows 10 PCs, right click the Windows Icon and type Settings
    2. In the Settings application, select Network & Internet, then select Change adapter options
    3. Using the mouse, right click Local Area Connection and select Properties
    4. Select the box next to Internet Protocol Version 4 (TCP/IPv4) (ensure that there is a check mark in the box)
    5. Select Internet Protocol Version 4 (TCP/IPv4) so that it is highlighted in blue then select Properties
    6. Select Use the following IP address and type 192.168.1.127
    7. Type 255.255.255 in the Subnet masks field
    8. No values are required for the other fields (leave blank) (Figure. 1)

      Figure 1. Internet Protocol Version 4 (TCP/IPv4) Properties

    9. Select OK and close out the settings
  3. Set up the static IP address on the Instrument
    1. The ICP-OES uses a default IP address of 192.168.1.128
    2. Open an Internet Explorer browser and type the default IP address (192.168.1.128) into the Web address bar. This should open the Agilent ICP-OES Web Utility welcome page (Fig. 2) 

      Figure 2. Agilent ICP-OES Web Utility welcome page

       

    3. The welcome page will show the ICP-OES instrument serial number. Verify that this value matches the instrument connected to your PC via LAN cable
    4. The Hostname and Description fields are set by user preference. The IP Address and Network Mask must match the Instrument Default IP address and the Subnet Mask entered into the PC settings. 
  4. Close the Web browser, then open Expert software to connect to the instrument in the software (Fig. 3)
    1. Select the Instrument page in ICP Expert, then select Connect and the Hostname or IP address of your instrument

      Figure 3.  Agilent ICP expert Instrument connection page

       

  5. If the values do not match the default values listed above, edit the values in the ICP-OES Web utility page, and select Apply to save these settings
    1. If a user name and password are required, the default user name is admin and the default password is Password. Select OK.
    2. If the Web utility requests restarting the instrument, press the button on the lower left front of the instrument (Standby button), and wait until the LED status indicator is no longer blinking, then press the power button again to restart the instrument.
    3. Proceed to step 4 above and connect the instrument in ICP Expert 
      1. If any changes were made to the default settings, select New Connection and enter the IP address that was saved in the Web utility page
  6. Successful connection between the instrument and the software can be verified by a solid green status LED on the front-right side of the instrument. 
  Tip
Learn how to effectively operate your Agilent ICP-OES:
ICPOES-5110-1200e - Agilent 5110 ICP-OES Instrumentation Overview
ICPOES-5900-1200e - Agilent 5800/5900 ICP-OES with ICP Expert: Instrumentation Overview e-learning courses available from Agilent education
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!