Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Typical Plasma Exhaust Temperature for 5000 ICP-OES Systems
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  26 Jul 2023
  • Modified :  2 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP OES Systems
  • Product Name :  5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES
  • Task :  Operation Site Preparation
Related Articles
  • Exhaust Requirements for the 5000 Series ICP-OES

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) Issue Correct ventilation is required to ensure proper function of the ICP-OES Resolution The ICP-OES must be located under a flue that is vented by an exhaust fan and ducted to an external vent. Caution : The exhaust system installation must comply with any rules and/or regulations that may be imposed by the local authorities responsible for control of facilities and fixtures in the work place...
  • ECD Baseline Troubleshooting

    This Information Applies To: All Agilent GC systems with, µECD detector Issue: The micro Electron Capture Detector (µECD) shows a higher than usual baseline signal. Warning : The ECD utilizes a radioactive source. You may not open the Electron Capture Detector (ECD) cell unless authorized to do so by your local nuclear regulatory agency. Do not disturb the four socket-head bolts. These bolts hold the cell halves together. Never eat, drink, or smoke when handling...
  • Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint

    This Information Applies To: Agilent GC systems, focusing on 8890 and 7890A/B, but also applies to 6850A/B, 6890A/N, 7820A, and 8860. Issue: If the inlet pressure / carrier gas flow fail to achieve the setpoint after a few minutes, the inlet pressure will shut down automatically. This article will help you to identify the possible causes of this issue. Steps to follow: Leak at the inlet To detect big leaks in the inlet, perform one of the diagnostic tests for the inlet: Inlet Leak Check...
  • Troubleshooting Agilent 5975 and 5977 GC/MS Air Leaks

    This Information Applies To: 5975 and 5977 GC/MS systems with MassHunter Acquisition software Issue: High-quality GC/MS data requires a low, and stable background of air and water present in the MS analyzer. This article demonstrates a systematic procedure to diagnose and troubleshoot a higher than normal air background. Introduction: Air and Water background will always be present in the GC/MS vacuum system. Air background is at m/z 28 (Nitrogen) and m/z 32 (Oxygen). The ratio of...
  • Software Status Indicators 5000 Series ICP-OES

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) Issue The ICP Expert software can help to identify system problems by assessing the status of the instrument components. Tip: Hardware LED status indicators are discussed in the article Resolution There are two LED displays that indicate instrument status: The AVS switching valve indicator light on the bottom right of the main software window shows: Green: the valve is in the inject...

Typical Plasma Exhaust Temperature for 5000 ICP-OES Systems

Answer

This Information Applies To: Agilent 5000 series ICP-OES


Issue

Exhaust temperature and requirements for laboratory site preparation


Background

Certain parameters will impact the actual temperature of the plasma exhaust. Some of these factors are: 

  • Ventilation
  • Air condition
  • Gas flows
  • RF power
  • Ambient temperature
  • Chiller set temperature
  • Filter conditions

 

Resolution

The typical plasma exhaust temperature is around 55 to 80 °C (130 to 180 °F).

The instrument limit is set to < 95 °C/200 °F.

Heat, ozone, vapors, and fumes generated by the plasma can be hazardous and must be extracted from the instrument by an exhaust system. Ensure that an exhaust system of the appropriate type is fitted (see Agilent 5100 and 5110 ICP-OES Site Preparation Guide).

The system must be vented to the outside air in accordance with local regulations and never within the building.

Regularly check the exhaust system by smoke test to ensure that the exhaust system is working correctly.

The exhaust fan must be always switched on before igniting the plasma.

The exhaust ducting should be corrosion resistant and fire-proof.

Due to the corrosive nature of some analytical work, it is recommended that in applications demanding high use of corrosive materials, an external cooling air supply system be provided.

It is recommended that the cooling air is supplied from an environmentally controlled area that is away from the instrument exhaust and any other area where corrosive materials are stored or used. The external cooling air system with flue, fan, and ducting must provide a minimum positive flow of 4 m3/minute (141 ft3/min) to the instrument when using the External Inlet Duct Adaptor Kit (G8010-68002).

 

 
Learn how to effectively operate your Agilent ICP-OES system:
CPOES-0GEN-1000e - Agilent ICP-OES Essentials: Introduction to Atomic Spectroscopy e-learning course available from Agilent education

 

Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Privacy Statement
Terms of Use
Contact Us
Site Help