Hi, Here's a screenshot of the system suit, there is no blank subtraction The analyst said he has used the same processing method for a different run with no issues the only difference being that for those runs the MS setting was the same time as the…
When you say that there are no signs of a leak, have you run a pump leak rate test using Agilent Lab Advisor? If not, I would suggest testing this. If you're unlucky enough for your peak to elute just at the point when the outlet ball valve fails to close…
The basic tools I would recommend to have for GC maintenance:
Ceramic column cutter
1/4" wrench (column nuts)
7/16" wrench (gas fittings at back of GC)
1/2" wrench (replace gold seal on S/SL inlet)
Torx 20 driver (removing FID collector assembly…
Hello and Help!
My GC-FID's inlet pressure stays at 0.7 psi and does not reach the setpoint of 10 psi. Makeup gas reaches the 25 psi setting. Liner and septa are fine. I unscrewed the inlet column nut and placed my finger at the bottom of the inlet…
Agilent GC 7890A, FID, Split Liner. I'm setting up the column to do Hexane injection testing. It was set up this way previously but the program had been deleted and the column removed before I started here. Front Detector/Column which is below baseline…
Hello,
My laboratory runs a multilevel calibration with internal standards on a headspace-GC-FID. Our sequence is set up to replace each level of the curve after each calibrator is analyzed and then calculate the concentration of the analyte with the…
I'm going to be calibrating for ethanol on my HS-GC-FID 6890, but I'm having discrepancies between two standards. The standard that I will be calibrating with (Restek's Residual Solvents #1 34105) is 3000ug/mL in N,N-Dimethylacetamide. I'm verifying the…
Hi Dr. Know
Here is some of the information you have requested. Please let me know if you need any further info.
How is your system configured? FID detector using helium as carrier gas, nitrogen as makeup gas.
Which phases are the two columns…
Hi
I am pretty new to GCMS, so may need extra details on how to do things you suggest
I attached a picture of the set up of our gc system, so I can explain better. We recently installed a guard column using 30 cm piece of FS Deactivated 0.250 mm(Agilent…
This Information Applies To: Agilent OpenLab ChemStation Edition Software (C.01.xx)
Issue
When your Agilent OpenLab ChemStation Edition software fails to launch with errors " Login failed with the following message: Configuration system failed to...
This Information Applies To: Agilent OpenLab ChemStation Edition Software (C.01.xx)
Issue
Agilent OpenLab ChemStation Edition software fails to launch with error containing the words " Autostart Macro failed ”. This article will provide you some...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer and Oligo Pro II instruments, and their respective controller software.
Issue Failed to build pressure error comes up shortly after starting a separation...
This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems equipped with a gate valve.
Issue
Removing the Ion Transfer Capillary for maintenance on instruments with gate valve. The gate valve allows ion capillary removal and...
This Information Applies To: Agilent OpenLab ChemStation, Agilent OpenLab CDS, Agilent OpenLab EZChrom
Issue
Due to the license error “ An error occurred connecting to the license server: The desired vendor daemon is down. Check the Imgrd log file...
This Information Applies To: HPLC Advisor app
Issue
Legacy HPLC method needs to be translated to a new column or LC system (HPLC to UHPLC or Analytical to Preparative LC).
Requirements
Install the most recent version of the HPLC Advisor...
This Information Applies To: Agilent 7890 A/B, 8860 and 8890 GC Systems with Split/Splitless (SSL), Multimode (MMI) or Programmable Temperature Vaporization (PTV) Inlets
Issue:
The Inlet Leak & Restriction Test is executed after every inlet maintenance...
This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7, ICP Expert version 7.X
Issue
The AVS can contribute to instrument performance issues such as sensitivity and precision. During troubleshooting, bypassing (or skipping...
This Information Applies To: Agilent GC Systems
Issue
Loss of peak resolution can be a combination of decrease in separation and increased peak width. Loss of resolution can impact on analytical reproducibility and accuracy.
Other chromatographic...
This Information Applies To: Agilent GC Systems
Issue
Contamination that creates ghost and carry over peaks can be introduced at different points including sample preparation, introduction, and separation. The additional peaks can cause false positive...