Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC Portal
  • Agilent Knowledge Portal
GC Portal
Articles Source of Ghost and Carryover Peaks in GC
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC Portal Home
  • GC Portal Articles
  • GC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC Sample Preparation and Introduction GC Systems
  • Component :  Automatic Liquid Samplers GC Inlets GC System GC Detectors Headspace Samplers PAL Autosampler Systems
  • Product Name :  6850 Automatic Liquid Sampler 6850 GC System 6890 GC System 7650A Automatic Liquid Sampler 7683 Automatic Liquid Sampler 7693A Automatic Liquid Sampler 7694 Headspace Sampler 7697A Headspace Sampler 7820 GC System 7890 GC System 8697 Headspace Sampler 8860 GC System 8890 GC System G1888A Headspace Sampler Intuvo 9000 GC System PAL 3 CTC Autosampler System PAL XT CTC Autosampler System
  • Task :  Troubleshooting
Related Articles
  • Excessive Background Noise in GC

    This Information Applies To: Agilent GC Systems Issue Excessive baseline noise can be introduced at different points including sample preparation, sample introduction, separation, and the detector. The excessive background can cause reduced signal-to-noise (detection limit), creating inaccurate quantitative or qualitative results. Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument in accordance with the recommended...
  • General Column Regeneration Procedure

    This Information Applies To: Agilent LC Systems Issue After analyzing dirty or concentrated samples, some sample compounds may retain in the column and other parts of the HPLC system. If the column re-equilibration after each run is not enough to elute the retained compounds, you will notice several performance problems. Tailing or wider peaks, higher background signal, carry over, or ghost peaks can be a sign of a contaminated column. The use of low-quality solvents can also be the cause...
  • Baseline Instability in GC

    This Information Applies To: Agilent GC Systems Issue Baseline instability that is not reproducible across multiple runs can be introduced at different points including sample introduction, separation, and the detector. The baseline instability causes integration problems, which can lead to inaccurate quantitative or qualitative results. Baseline rises reproducibly, are normal for ramped temperature programs. The change in baseline will be consistent across multiple runs and multiple days...
  • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) and Agilent 700 Series ICP-OES (710/715/720/725/730/735) Issue Loss or change of sensitivity in ICP-OES can occur under normal operating practices. This discussion identifies factors that can contribute to sensitivity changes including calibration standards , memory effects , and the sample introduction system . Sample introduction components include: Autosampler probe Switching valve Pump tubing Nebulizer...
  • Recommended Routine ICP-OES and AVS 6/7 Maintenance Schedule

    This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900) Issue: The ICP-OES and AVS 6/7 components require regular routine maintenance. Agilent-authorized field service engineers must perform any maintenance procedures not listed in this document. Warning : Shock Hazard. Do not remove instrument covers that expose electrical circuits, devices, and components operating at dangerous voltages. Contact with these circuits, devices, and components can result in...

Source of Ghost and Carryover Peaks in GC

Answer

This Information Applies To: Agilent GC Systems


Issue

Contamination that creates ghost and carry over peaks can be introduced at different points including sample preparation, introduction, and separation. The additional peaks can cause false positive identifications and/or inaccurate quantitative results. 

Other chromatographic problems are identified in Basic Troubleshooting for GC Systems. Regular maintenance of your instrument in accordance with the recommended maintenance schedule will reduce the incidence of chromatography problems.


Background

Over time sample deposits can accumulate in the analytical pathway and slowly leach out during subsequent sample analyses. This situation can cause peaks that are not from the current sample to appear in your chromatogram (Figure 1). The additional peaks can appear at any random retention time over one or more analytical runs and include:

  • Ghost peaks: Unknown, symmetrical, or unsymmetrical peaks that can occur at any time that are not expected to come from the current sample
  • Carryover peaks: Peaks from previous injections.


Figure 1. Ghost peaks and carryover
1. Normal contamination free run, 2. Contaminated with ghost peaks and carry over

Resolution

Possible sources of ghost peaks and carryover include:

  1. Sample preparation

    Extra peaks can be present due to sample preparation:

    Possible source of contamination Solution
    Low quality or contaminated solvent Check to see if the problems coincided with a change in solvent supply. If they did, change to a known contamination free solvent supply.
    Sample contains contaminants Sample or extract clean up before the analysis.
    Sample contaminated before introduction to the GC Check sample handling steps. Repeat sample preparation as necessary. Possible sources: Sample cleanup, handling, transfer, and storage.

     

  2. Sample introduction system 

    Perform the GC Condensation Test. If the condensation test determines the sample introduction system is the source of the contamination, continue with the following:

    Possible source of contamination Solution
    Low quality or contaminated gas supply Check to see if the problems coincided with a change in gas cylinder. If they did, change the gas cylinder and flush the gas lines for sufficient time to clear the volume of the gas system.
    Contaminated gas filter Check the gas filter indicator (if available). If not, replace as per the maintenance schedule or after using contaminated gas.
    Syringe contamination Change the syringe for a known clean one. If the problem does not recur, the syringe was dirty
    Septum bleed or degradation Replace the septum. Use an Agilent high-quality septum appropriate for your inlet temperature. See Replacing the Inlet Septum for Split/Spitless and Multimode Inlets 
    Inlet liner dirty Replace the inlet liner. Use an Agilent inlet liner appropriate for your injection volume, solvent, and injection mode
    Gold seal dirty Replace the gold seal. See How to Replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC 
    Inlet contaminated Clean or Bake out the inlet. See How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems 
    Split vent line contaminated or blocked Clean or replace the split vent line and the split trap. Review your method to ensure that inlet liner volume is sufficient for your injection volume.  

     

  3. Analytical column

    Possible source of contamination Solution
    Column contamination Bake out the column. Limit the bake-out for 1 to 2 hours
    Caution 
    Caution: Do not exceed the manufacturer recommended maximum column temperate limit as it will damage your column.
    Trim the column: Remove 0.5 to 1 m from the front of the column or 

If these procedures have not resolved the ghost peak/carryover problem, contact Agilent Technical Support.

 Tip 
Learn how to effectively troubleshoot your Agilent GC System:
GC-0GEN-1040z - Practical Steps in GC Troubleshooting
GC-0GEN-2040zs - GC Troubleshooting Series e-learning courses available from Agilent education

 

 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help