This Information Applies To: Agilent GC Systems
Issue
Contamination that creates ghost and carry over peaks can be introduced at different points including sample preparation, introduction, and separation. The additional peaks can cause false positive identifications and/or inaccurate quantitative results.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems. Regular maintenance of your instrument in accordance with the recommended maintenance schedule will reduce the incidence of chromatography problems.
Background
Over time sample deposits can accumulate in the analytical pathway and slowly leach out during subsequent sample analyses. This situation can cause peaks that are not from the current sample to appear in your chromatogram (Figure 1). The additional peaks can appear at any random retention time over one or more analytical runs and include:
Figure 1. Ghost peaks and carryover
1. Normal contamination free run, 2. Contaminated with ghost peaks and carry over
Resolution
Possible sources of ghost peaks and carryover include:
Sample preparation
Extra peaks can be present due to sample preparation:
Possible source of contamination | Solution |
Low quality or contaminated solvent | Check to see if the problems coincided with a change in solvent supply. If they did, change to a known contamination free solvent supply. |
Sample contains contaminants | Sample or extract clean up before the analysis. |
Sample contaminated before introduction to the GC | Check sample handling steps. Repeat sample preparation as necessary. Possible sources: Sample cleanup, handling, transfer, and storage. |
Sample introduction system
Perform the GC Condensation Test. If the condensation test determines the sample introduction system is the source of the contamination, continue with the following:
Possible source of contamination | Solution |
Low quality or contaminated gas supply | Check to see if the problems coincided with a change in gas cylinder. If they did, change the gas cylinder and flush the gas lines for sufficient time to clear the volume of the gas system. |
Contaminated gas filter | Check the gas filter indicator (if available). If not, replace as per the maintenance schedule or after using contaminated gas. |
Syringe contamination | Change the syringe for a known clean one. If the problem does not recur, the syringe was dirty |
Septum bleed or degradation | Replace the septum. Use an Agilent high-quality septum appropriate for your inlet temperature. See Replacing the Inlet Septum for Split/Spitless and Multimode Inlets |
Inlet liner dirty | Replace the inlet liner. Use an Agilent inlet liner appropriate for your injection volume, solvent, and injection mode |
Gold seal dirty | Replace the gold seal. See How to Replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC |
Inlet contaminated | Clean or Bake out the inlet. See How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems |
Split vent line contaminated or blocked | Clean or replace the split vent line and the split trap. Review your method to ensure that inlet liner volume is sufficient for your injection volume. |
Analytical column
Possible source of contamination | Solution | ||
Column contamination | Bake out the column. Limit the bake-out for 1 to 2 hours
|
If these procedures have not resolved the ghost peak/carryover problem, contact Agilent Technical Support.