Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Articles Failed to Build Pressure Error on a Parallel Capillary Electrophoresis Instrument
  • Articles
  • Files
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Automated Electrophoresis Portal Home
  • Automated Electrophoresis Portal Articles
  • Automated Electrophoresis Portal Files
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  7 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Automated Electrophoresis Systems
  • Component :  Femto Pulse Systems Fragment Analyzer Systems Oligo Pro II Systems ZAG DNA Analyzer Systems
  • Product Name :  5200 Fragment Analyzer System 5300 Fragment Analyzer System 5400 Fragment Analyzer System Femto Pulse System Oligo Pro II System ZAG DNA Analyzer system
  • Task :  Troubleshooting
Related Articles
  • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, Oligo Pro II Issue: This article explains the steps to take after installing a brand new array in the Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, and Oligo Pro II systems. Steps to follow Capillary array replacement steps are detailed in chapter 8 of the Fragment Analyzer user manual , Femto Pulse user manual , ZAG DNA analyzer user manual , and chapter 9 of the Oligo Pro II user manual...
  • Agilent Parallel Capillary Electrophoresis Compatible Plasticware

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, and ZAG Issue: Only plasticware from the approved consumables list should be used in the Agilent Parallel Capillary Electrophoresis systems. The use of plasticware with different dimensions could lead to decreased injection quality and consistency as well as damage to the tips of the instrument capillaries. Resolution: Compatible plates and tubes are listed in the following manuals: 5200, 5300...
  • Insert Disk Error in the Agilent Fragment Analyzer Controller Software

    This Information Applies To: 5200/5300/5400 Fragment Analyzer controller software, Femto Pulse controller software, ZAG DNA Analyzer controller software, Oligo Pro II controller software Issue: The Agilent Fragment Analyzer, Femto Pulse, ZAG, or Oligo Pro II controller software does not open when the desktop icon is double-clicked. Instead, one of the following errors come up: "Insert the 'Fragment Analyzer' disk and click OK" "The feature you are trying to use is on a network resource...
  • Long-Term Capillary Array Storage on Parallel Capillary Electrophoresis Instruments

    This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, and Oligo Pro II Issue Methods for storing capillary array when not in use for longer than two weeks. The array can either be stored installed in the instrument or in the external array docking station. Items Required Capillary Storage Solution For use with the Femto Pulse and Oligo Pro II system (p/n GP-435-0100 ) For use with the Fragment Analyzer and ZAG DNA Analyzer systems...
  • Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, ZAG, and their respective controller software Issue: A full power cycle can help to recover communication/connection issues of Agilent parallel capillary electrophoresis instruments (5200/5300/5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, and ZAG systems). Communication and connection issues can materialize already during software start as connection failed error messages although the instrument is...

Failed to Build Pressure Error on a Parallel Capillary Electrophoresis Instrument

Answer

This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer and Oligo Pro II instruments, and their respective controller software.


Issue
Failed to build pressure error comes up shortly after starting a separation.


Resolution

Failure to build pressure error occurs when the system is trying to build pressure when liquid is being pumped during conditioning or flushing of the instrument. Possible causes and suggestions to address the issue are: 

1. Empty reagent bottles - As first step, please ensure that there is adequate volume in each reagent bottle and that the respective tubing is in the liquid.

2. Leak - Please check whether there is any visible leak at the tubing (any visible liquid inside the instrument). Figure 1 shows various red tube fittings and other components of the side compartment of the fragment analyzer instrument. The points are going to be similar for other instruments as well. Check for leaks in these areas including but not limited to what is visible in the picture. 

  • If there is any obvious and easily accessible leak, for example around the array bundle, reinstall it into the reservoir (Figure 1, (1)). If there is liquid at the F port (Figure 1, (4)) try to fix by tightening the red tube fittings. 
  • If leak is from the tube that connects to the reservoir, then take a picture (Figure 1) and contact your local Agilent Support from Contact us page. This connection is not usually something users can tighten themselves so an onsite visit by Agilent field engineer will likely be needed.  
  • If there is no leak and issue is not leak related, please proceed to step 3.


Figure 1. Fragment Analyzer side door compartment.


Various leak points are: (1) Top of the array bundle (2) Vent valve (3) F-port connections, located by red ferrules (4) Conditioning line (5) syringe. 

3. Reboot - Perform a full system reboot following, see the steps in the article: Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments. Full power cycle is recommended to address any possible component disconnects. If that resolves the issue, it is recommended to monitor for some time. If not, please proceed to step 4.

4. Pump related - To determine whether issue is pump-related, the recommendation is to execute a Conditioning Solution flush so that pump moves few times. Click Add to queue, select Full Conditioning or any method using Conditioning Solution and edit it to only contain the Conditioning step. Click OK twice to add it to the queue. (Figure 2).  During the flush step, please monitor syringe and liquid movement. 

    

Figure 2. Adding Capillary Array Conditioning to queue. 

  • Syringe movement indicates that the pump is working.
  • If pump does not move, then the issue is likely instrument hardware failure. An onsite visit by an Agilent engineer will be needed. Therefore, please collect and share instrument Event and Error reports (Figure 3) from the instrument controller software (Admin > Error or Event Report > choose at least dates at which the error occurred) with your local Agilent support from Contact us page. 


Figure 3.  Steps to collect Event and Error reports from controller software

 

 Tip 
Learn how to effectively use your parallel capillary electrophoresis instruments:
5200, 5300 and 5400 Fragment Analyzer User Manual 
Femto Pulse User Manual 
ZAG DNA Analyzer User Manual 
Oligo Pro II User Manual

For Research Use only. Not for use in diagnostic procedures.

PR 7000-8759

 