Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC/MS Portal
  • Agilent Knowledge Portal
LC/MS Portal
Articles Removing the Ion Transfer Capillary from Triple Quadrupole LC/MS Systems Equipped With a Gate Valve
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC/MS Portal Home
  • LC/MS Portal Articles
  • LC/MS Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  LC/MS Systems
  • Component :  LC/MS Triple Quadrupole
  • Product Name :  6475 Triple Quadripole LC/MS 6495 Triple Quadripole LC/MS Ultivo Triple Quadrupole LC/MS
  • Task :  Maintenance
Related Articles
  • Installing an Ion Capillary in Triple Quadrupole LC/MS Systems Equipped With a Gate Valve

    This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems equipped with a gate valve Issue Installing the Ion Transfer Capillary after cleaning or replacement in Agilent LC/MS instruments with a gate valve. Required Items Clean powder-free nitrile gloves LC/MS-grade isopropanol Caution : Ensure the source heated zones are cool before working on the instrument. Before installing the new ion capillary, the used ion capillary must be removed...
  • What to do if my LC-QQQ Autotune Fails?

    This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems Issue Autotune fails to complete. Background Agilent 6400 LC/MS systems use an Autotune algorithm to optimize the sensitivity and mass resolution of the instrument. Sometimes when running the autotune procedure the autotune does not complete and the user is given an Autotune failed message in MassHunter. One of the main reasons for Autotune failure is a lack of sensitivity or abundance of the tuning ions...
  • Removing the Ion Transfer Capillary from Agilent 6000 Series LC/MS Systems

    This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal. For instructions for removing an ion capillary on LC/MS instruments with a gate valve installed, please click here Issue: Removing the Ion Transfer Capillary for maintenance. Steps to follow: Tips : The figures in this article show an LC/MS with an Agilent Jet Stream (AJS) source installed. The process of capillary removal is the same for LC/MS instruments...
  • Installing the Ion Transfer Capillary in Agilent 6000 Series LC/MS Systems

    This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal. Issue: Installing the Ion Transfer Capillary after cleaning or replacement. For instructions for installing an ion capillary on LC/MS instruments with a gate valve installed, click here Steps to follow: Tips : The following figures show an LC/MS with an Agilent Jet Stream (AJS) source installed. The process of capillary installation is the same for...
  • How To Ping the IP Address of an Agilent GC System

    This Information Applies To: Agilent GC Systems 6890N, 7890A/B, 7820, 8860,8890, Intuvo 9000 Issue The communication between Agilent GC instrument and the computer (PC) requires an active LAN connectivity. Each part of the instrument-PC network has a unique IP Address, a shared Subnet Mask, and an undefined Gate Way. This ping test verifies the LAN connectivity between PC and instrument. If the pinging is not successful, the customer will not be able to configure the new instrument or the configured...

Removing the Ion Transfer Capillary from Triple Quadrupole LC/MS Systems Equipped With a Gate Valve

Answer

This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems equipped with a gate valve. 


Issue

Removing the Ion Transfer Capillary for maintenance on instruments with gate valve. The gate valve allows ion capillary removal and installation without having to vent the instrument.


Required Items

  • Clean powder-free nitrile gloves
  • Capillary puller tool (supplied with your triple quadrupole LC/MS with a gate valve installed)

Steps to follow

 Warning 
Caution: Handling Ion Source components or the Ion Transfer Capillary without gloves can cause contamination. Please use clean powder-free nitrile gloves during this procedure

1. Turn off heated zones in the instrument and wait for a least 20 minutes for the mass spectrometer's heated zones to cool down.

2. Put on clean powder-free nitrile gloves before opening the ion source.

 Warning 
Warning: The spray chamber operates a very high temperature. Do not open the spray chamber until it is cool.

3. Open the spray chamber.

4. Carefully test that the spray shield is cool to the touch then remove the spray shield and capillary end cap as shown in Figure 1.

 
Figure 1 Removing the spray shield and end cap from the desolvation assembly of the mass spectrometer.

 
Figure 2. Arrow showing open collet on the capillary puller tool.

5. Screw the capillary puller tool fully into the spray shield mount using the installation knob. Then push collet assembly over capillary and tighten by holding the puller handle and turning the collet clamping knob clockwise. See Figure 3.


Figure 3. Installing the capillary puller tool on the desolvation assembly and tightening the collet on the end of the capillary in preparation of removal.

6. Pull on the puller handle until the collet assembly is fully retracted and stops moving. Visually check by looking through the slots in the tool body that the capillary has been pulled out by the collet. See Figure 4.

7. Slide the lock button of the gate valve actuator lever towards the back of the instrument with one hand (finger). With your other hand, lift the actuator lever until it fully stops. Release the lock button. See Figure 4.

 
Figure 4. Fully retracting the collet assembly and then closing the gate valve using the actuator lever.

8. Unscrew the tool from the spray shield mount. Carefully pull the tool straight out to avoid applying any bending forces to the capillary. See Figure 5.


Figure 5. Unscrewing the puller tool from the spray shield mount and pulling the capillary straight out of the desolvation assembly.

 Warning 
Caution: Perform the next step over a clean bench or other working surface to reduce the chance of breaking a capillary by dropping it from the collet.

9. Hold the puller handle and rotate the collet clamping knob two to three turns counterclockwise (anticlockwise) to loosen the collet. Push the collet clamping knob while holding the puller handle to free the capillary from the collet. See Figure 6.


Figure 6. Removing the capillary from the capillary puller tool.

For instructions on cleaning the ion capillary see How to Clean the Ion Capillary in Agilent 6000 Series LC/MS Instruments . For instruction on re-installing your ion capillary, see Installing an Ion Capillary in Triple Quadrupole LC/MS Systems Equipped With a Gate Valve .

 Tip 
Learn how to effectively maintain your Agilent LC/MS:
LCMS-6400-1201s - Essential Housekeeping Tasks and Best Practices for Agilent 6400 Series LC/TQ e-learning course available from Agilent education
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help