Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
Greetings Agilent Community,
As with many of the errors I encounter with my system, these two happened after I foolishly left the lab to pursue a life outside the world of LCMS.
(1) From the instrument log, the first error read "Failed to run acquisition…
When attempting to submit a sequence in EZChrom, it fails to submit and returns this error message:
This happens each time a sequence is submitted. Any help would be appreciated!
Hi Agilent community.
I'm a LC-MS equipment user and I had a problem with the MassHunter Acquisition B 08.01 software.
When I try to open the Acquisition software, appears the software logo but immediately appears an error dialog box. This error dialog…
Hi everyone,
We have a 6546 QTOF, the following error keeps popping up during acquisition; Failed to start logbook server. A shared violation has occurred. How can I resolve this?. I'll appreciate your help. Thanks
This Information Applies To: Agilent OpenLab ChemStation Edition Software (C.01.xx)
Issue
When your Agilent OpenLab ChemStation Edition software fails to launch with errors " Login failed with the following message: Configuration system failed to...
This Information Applies To: Agilent OpenLab ChemStation Edition Software (C.01.xx)
Issue
Agilent OpenLab ChemStation Edition software fails to launch with error containing the words " Autostart Macro failed ”. This article will provide you some...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer and Oligo Pro II instruments, and their respective controller software.
Issue Failed to build pressure error comes up shortly after starting a separation...
This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems equipped with a gate valve.
Issue
Removing the Ion Transfer Capillary for maintenance on instruments with gate valve. The gate valve allows ion capillary removal and...
This Information Applies To: Agilent OpenLab ChemStation, Agilent OpenLab CDS, Agilent OpenLab EZChrom
Issue
Due to the license error “ An error occurred connecting to the license server: The desired vendor daemon is down. Check the Imgrd log file...
This Information Applies To: HPLC Advisor app
Issue
Legacy HPLC method needs to be translated to a new column or LC system (HPLC to UHPLC or Analytical to Preparative LC).
Requirements
Install the most recent version of the HPLC Advisor...
This Information Applies To: Agilent 7890 A/B, 8860 and 8890 GC Systems with Split/Splitless (SSL), Multimode (MMI) or Programmable Temperature Vaporization (PTV) Inlets
Issue:
The Inlet Leak & Restriction Test is executed after every inlet maintenance...
This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7, ICP Expert version 7.X
Issue
The AVS can contribute to instrument performance issues such as sensitivity and precision. During troubleshooting, bypassing (or skipping...
This Information Applies To: Agilent GC Systems
Issue
Loss of peak resolution can be a combination of decrease in separation and increased peak width. Loss of resolution can impact on analytical reproducibility and accuracy.
Other chromatographic...
This Information Applies To: Agilent GC Systems
Issue
Contamination that creates ghost and carry over peaks can be introduced at different points including sample preparation, introduction, and separation. The additional peaks can cause false positive...