Hello,
Having sorted out our printer problem with MassHunter for GC we started having another issue on Friday night last week. The software suddenly stopped mid run giving an error that the drivers for the MSD could not be found. After restarting…
Hi. The pump made loud noises for several days. After that, the pump stopped working normally. Near 600 rpm the pump stops and A-PLUS Software was showing fail status (205) with error code 206 (bit 128) - too hight load. We use Agilent pump - TPS compact…
Hello,
We have been having issues with our ICP MS performance and tune reports. Both have been failing with low counts. We have changed the peri-pump tubing, remade all standards, remade the tune solution, replaced the torch, and cleaned the cones…
Hi,
Recently Run the Wavelength accuracy and signal to noise at FLD detector for qualification purposes but it not passing. In wavelength accuracy test got Excitation value 350 it ok but in case of emission got 393 nm its failed required minimum 394…
Thanks for advising, do other errors show up in the logbook?
Please watch the movement of the sampler arm during initialisation:
- Is it shaky or jittery? If yes, turn off module and clean transport rods using IPA & lint free cloth.
- Which position…
Hi Everyone, I am hoping I could get some advice with separating Aniline and DMSO with GC-FID.
We tried DB-Wax, DB-624, DB-5 so far but it did not work well. The retention time could be 2 min different max but the problem is DMSO has some impurities…
Hello, I have an issue when I try to install OpenLab CDS EZChrom Version A.04.07 The plan is to install EZChrom on a Citrix to allow the user to retrieve files from ECM for review or reintegration. I start the installation and select the Installation…
Hi I had a question about GC MS Split (gc to FID/MS). I have a thermo splitter that has no auxillary flow to counteract the ms vaccuum, but want to change out the capillaries that come recommended with the splitter. The recommended capillary dimensions…
Hi ronin, It sounds like you have a 7694A with the newer headspace mainboard (7694B) that didn't have the visible changeable battery. In this case the ability of the board to retain stored parameters has failed due to this battery failing. The only fix…
This file contains a list of 14 commonly used consumables and parts for Intuvo 9000 GC .
Upload this file to your Smart Alerts software to automatically add these parts to the Master Parts List
This Information Applies To: Agilent MassHunter 11 and later
Issue
Generating Acquisition Desktop and Quantitative Analysis shortcuts.
Steps to Follow
Double-click the Control Panel icon on the desktop or select Windows Start menu > Agilent...
This Information Applies To: Agilent MassHunter Workstation for LC/TQ and MassHunter Workstation for LC/TOF and LC/Q-TOF
Issue
How to view the MassHunter Data Acquisition Software Logs entries, change settings, and export the Logbook for archiving...
This Information Applies To: ICP-MS MassHunter with User Access Control (UAC)
Issue
It is necessary to update the database folder path in Change User Access Control:
After creating a new database in SDA with ICP-MS MassHunter.
To view a batch...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
Over time, the electrodes may become pitted or eroded and users may see a decrease in precision or begin to experience the "Tube fault" error when new tubes are installed. If...
This Information Applies To: Agilent GTA Furnace and Flame Instruments, Agilent 240FS/240GFAA/GTA120/PSD120
Issue
There may be occasions when users need to remove GTA furnace center block workhead (instruments with both Flame and Furnace capabilities...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
The shroud should be replaced annually, or when there is evidence of damage caused by mechanical shock or contamination from liquid spillage.
Items Required
Shroud Removal...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
Before replacing the electrodes and/or shroud, the GTA Furnace workhead will need to be disassembled. After replacing the electrodes and shroud, the GTA Furnace workhead will...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer
Issue
After starting a run, the run is aborted with the following error: Current = 0.0uA, Expected range: 5 - 3000.
Steps to Follow
Check...
This Information Applies To: Agilent OpenLab CDS 2.x in file system configuration
Issue
Sometimes, you will need to change the Projects Root Path / Project Storage Path . This article outlines the procedure on how to change it.
Background
...
This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7
Issue
Steps to bypass the AVS to troubleshoot poor performance on the ICP-OES
Tip : These steps should be completed after the Subsystem Communication...