Hello ivynotterpek ,
I'm not aware of any way to do this with just one compound/cal curve. One solution would be to make duplicate compounds and assign one compound the low level standards and the other the high level standards. You can utilize different…
Thank you, unfortunately we tried to set a collection window much larger than the peak duration, but the peak got split again. And in the version of the software we are using no advanced settings allowing to collect different peaks into the same vial…
Hi CMZainea,
Thank you for your feedback. The leak test for the 6 ports valve was passing, so no leaks detected at the valve. Any other leaks were detected from the visible piping and connection from the back panel of the 7697A going to the 6 ways valve…
The two-way splitter without makeup G3181B can be used as a splitter or a combiner to connect two inlets to one detector. Discuss this part with your local sales rep and they can help you get a quote for this part.
https://www.agilent.com/cs/library…
The Digest of upcoming webinars has been updated as of February 20.
Notable additions include:
BioPharma (1)
GC & GCMS (2)
Pharma (1)
Software & Informatics (21)
Check Out:
Tips and Tricks Workshop on ICP-OES: Smarter Method Development…
This Information Applies To: Agilent GC Systems
Issue
This test isolates the point of contamination to the sample introduction system, which includes the gas supply, gas supply lines, and the sample inlet. It does not include the syringe or autosampler...
This Information Applies To: Agilent GC systems
Issue
Peak tailing is present in a chromatography peak when it has an excessive asymmetry with a trailing edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
T his Information Applies To: Agilent 8890/8860/7890/7820/6850/6890 GC Systems with TCD (Thermal Conductivity Detector)
Issue
Problems observed with thermal conductivity detector (TCD) results can have different underlying causes.
Resolution...
This Information Applies To: All GCs, 6890, 6850, 7890, 7820, 8890/8860, Intuvo 9000
Issue:
This article is a troubleshooting guide for when the Flame Ionization Detector (FID) does not ignite or the flame keeps going out.
Cause:
If the FID...
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue:
The Flame Ionization Detector (FID) jet could become clogged with ferrule material, column bleed, or heavy molecular analytes. The clogged jet will affect...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different high concentration, split analysis techniques including:
General use liners
Auto inject, general use
Auto...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different splitless trace analysis techniques including:
General use liners
Pressure, temperature, volume (PTV)
Hot...
This Information Applies To: Agilent 6890, 7890, 7820, 8890 and 8860 GC Systems with Split/Splitless (S/SL) inlet
Issue
Replacement of the Gold Seal is required as part of inlet maintenance. Replacement is required when the performance of the inlet...
This Information Applies To: Agilent GC Systems with Micro-ECD.
Issue
The Wipe Test is an annual test that performed by end user of the micro-ECD (Micro Electron Capture Detector or uECD) to help ensure radioactive safety when using the detector...
This Information Applies To: Agilent GC Systems with thermal conductivity detector (TCD)
Issue
Baking the thermal conductivity detector (TCD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.