Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC Portal
  • Agilent Knowledge Portal
GC Portal
Articles Peak Tailing in GC Systems
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC Portal Home
  • GC Portal Articles
  • GC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  1 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC Sample Preparation and Introduction GC Systems
  • Component :  Automatic Liquid Samplers GC Inlets GC System GC Detectors Headspace Samplers PAL Autosampler Systems
  • Product Name :  6850 Automatic Liquid Sampler 6850 GC System 6890 GC System 7650A Automatic Liquid Sampler 7683 Automatic Liquid Sampler 7693A Automatic Liquid Sampler 7694 Headspace Sampler 7697A Headspace Sampler 7820 GC System 7890 GC System 8697 Headspace Sampler 8860 GC System 8890 GC System G1888A Headspace Sampler Intuvo 9000 GC System PAL 3 CTC Autosampler System PAL XT CTC Autosampler System
  • Task :  Troubleshooting
Related Articles
  • Peak Fronting in GC

    This Information Applies To: Agilent GC Systems Issue Peak fronting is present in a chromatography peak when it has an excessive asymmetry with a leading edge. A normal peak is almost symmetrical. Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument in accordance with the recommended maintenance schedule will reduce the incidence of chromatography problems. Background Fronting can be caused by sample solvent,...
  • Modify the Integration Parameters in MassHunter Quantitative Analysis

    This Information Applies To: Agilent MassHunter Quantitative Analysis Software Issue Chromatographic peak integration is a critical part of accurate and reproducible quantitative data analysis. The integrator type and the integration parameters can be adjusted to optimize peak integration. Available Integrators MassHunter Quantitative Analysis has three categories of integrators, with a total of seven automated integrator types. The results from all integrators can be filtered with the...
  • Poor Precision in ICP-OES

    This Information Applies To: Agilent 5000 series ICP-OES systems Issue Poor or high relative standard deviation, or a noisy signal is an indication of poor precision during analysis. Resolution There are many potential causes for an instrument exhibiting poor precision, including but are not limited to: V-groove nebulizer Concentric glass nebulizer Torch Contamination in spray chamber Pump/pump tubing Air leak Instrument calibration Aqueous analysis after non-aqueous...
  • Replacing a Broken Transfer Line on 7697A Headspace Sampler

    This Information Applies To: Agilent 7697A Headspace Sampler Issue The transfer line will need to be replaced if it breaks at the valve, the inlet or in the middle of the transfer line. The previously broken liner will need to be removed and new one installed. You may experience leak-like symptoms or poor/no chromatography. Items required 3/16- inch open end wrench Column cutting wafer 1/4- inch open end wrench Wrench, angled, septum nut (GC inlet wrench) Polyimide ferrule corrected...
  • Removing and Installing a Syringe on the Agilent 7693A Autosampler

    This Information Applies To: Agilent 7693A Automatic Liquid Sampler Issue: The automatic liquid sampler may need to be replaced, in case it is worn or dirty. Steps to follow To remove the syringe: Open the injector door. Slide the syringe carriage to the top position if necessary. Completely loosen the plunger thumb screw until it reaches the stop, and lift the plunger carrier off of the syringe plunger. Open the syringe latch by swinging it in a counterclockwise...

Peak Tailing in GC Systems

Answer

This Information Applies To: Agilent GC systems


Issue

Peak tailing is present in a chromatography peak when it has an excessive asymmetry with a trailing edge. A normal peak is almost symmetrical.  

Other chromatographic problems are identified in  Basic Troubleshooting for GC Systems . Regular maintenance of your instrument in accordance with the recommended maintenance schedule will reduce the incidence of chromatography problems.


Background

Tailing can be caused by several reasons. Figure 1 shows some examples of peak tailing at several degrees. 

           
Figure 1. Peak tailing
1. Symmetrical peak with no tailing 

Resolution

 Tip 
Tip: Some active compounds always tail. 

 

Possible cause of tailing Solution
Split ratio too low  Increase the split ratio
Column contamination Bake out the column. Limit the bake-out for 1 to 2 hours or until the signal is stable
 Caution
Caution: Do not exceed the manufacturer recommended maximum column temperate limit as it will damage your column.
Trim the column: Remove 0.5 to 1 m from the front of the column 
Column activity Irreversible. Replace the column
Improper injection technique (slow/erratic plunger depression) Check that the syringe plunger is moving freely. The solution should be pulled into the syringe body, and moving freely and fast when plunger is depressed
Improper column installation, including poor column cutting Remove and reinstall the column. See Capillary column installation quick reference guide 
Solvent effect violation for splitless or on-column injections Lower the oven start temperature
Mixed sample solvent with large differences in polarity or boiling points Change the sample solvent or use a retention gap


If these procedures have not resolved the peak tailing problem, contact Agilent Technical Support.

 Tip 
Learn how to effectively troubleshoot your Agilent GC System:
GC-0GEN-1040z - Practical Steps in GC Troubleshooting
GC-0GEN-2040zs - GC Troubleshooting Series e-learning courses available from Agilent education
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help