Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC Portal
  • Agilent Knowledge Portal
GC Portal
Articles Using the Condensation Test to Troubleshoot the GC Inlet
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC Portal Home
  • GC Portal Articles
  • GC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC Systems
  • Component :  GC Inlets GC System
  • Product Name :  6850 GC System 6890 GC System 7820 GC System 7890 GC System 8860 GC System 8890 GC System Intuvo 9000 GC System
  • Task :  Troubleshooting
Related Articles
  • General Maintenance for the Agilent Flame and Graphite Furnace Atomic Absorption Systems

    This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace) Issue: This article includes recommended maintenance procedures to be performed daily, weekly or annually to ensure optimum performance of the Agilent atomic absorption systems. Steps to follow: Schedule of recommended maintenance procedures: Daily Check that the exhaust system has positive extraction Check the gas supplies Check...
  • Troubleshooting FID Noise and High Background

    This Information Applies To: All Agilent GC systems with flame ionization detector (FID) Issue: The FID uses a hydrogen flame to burn organic samples, resulting in increased ionization that is measured as a current from 5 to 10^6 pA. Normal FID background levels will be in the 5 to 20 pA range, with no sample present and the GC oven at low temperature. Agilent recommends a temperature ≥ 300 °C to prevent condensation damage. The detector temperature should be approximately 20 °C greater than...
  • Checking for a Blocked ICP-OES Nebulizer

    This Information Applies To: Agilent 5000 Series ICP-OES Issue Blocked nebulizers are the source of many instrument performance issues including loss of sensitivity and poor precision, especially when running organics, high solids, or high salt concentration solutions. The ICP-OES backpressure Nebulizer test verifies nebulizer performance. Run this test if you notice no mist in the spray chamber, have low sensitivity during analysis, or high % RSDs on your replicates. Blockages and partial...
  • How to Blank off a GC/MS Instrument for Troubleshooting

    This Information Applies To: Agilent single quadrupole and triple quadrupole GC/MS systems Issue During troubleshooting, it can be useful to isolate the MS detector from the rest of the GC system (sample introduction, inlet, column etc.). Background Blanking off the MS will help to determine if an issue is coming from the MS detector itself or the GC. Some examples where this procedure might be useful: Noisy or unusual baseline. Tuning difficulties. Leaks (high oxygen and nitrogen...
  • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) and Agilent 700 Series ICP-OES (710/715/720/725/730/735) Issue Loss or change of sensitivity in ICP-OES can occur under normal operating practices. This discussion identifies factors that can contribute to sensitivity changes including calibration standards , memory effects , and the sample introduction system . Sample introduction components include: Autosampler probe Switching valve Pump tubing Nebulizer...

Using the Condensation Test to Troubleshoot the GC Inlet

Answer

This Information Applies To: Agilent GC Systems 


Issue

This test isolates the point of contamination to the sample introduction system, which includes the gas supply, gas supply lines, and the sample inlet. It does not include the syringe or autosampler. 


Background

Typical chromatographic features that indicate possible contamination problems include ghost peaks, carryover, excessive baseline noise or baseline drift, and instability. See Basic Troubleshooting for GC Systems .

Steps to Follow

To identify if the inlet is the source of contamination: 

  1. Leave the GC between 40 to 50 °C for eight or more hours

  2. Run a blank analysis (i.e., start the GC, but with no injection) using the normal temperature conditions and instrument settings

  3. Collect the chromatogram for this blank run

  4. Immediately repeat the blank run when the first run one is completed. Do not allow more than 5 minutes to elapse before starting the second blank run

  5. Collect the chromatogram for the second blank run and compare it to the first chromatogram. If:

    1. Both chromatograms contain few peaks or little baseline drift, the carrier gas, carrier gas lines, gas filters and inlet are relatively clean

    2. The first chromatogram contains a larger number of peaks and baseline instability, any combination of the incoming carrier gas, carrier gas line, gas filters or the inlet is contaminated. See Basic Troubleshooting for GC Systems .

       Tip 
      Tip: Gas supply problems are most likely to occur immediately after changing gas cylinders. 

       

 Tip 
Learn how to effectively troubleshoot your Agilent GC System:
GC-0GEN-1001e - Gas Management Best Practices
GC-0GEN-1012s - GC Inlets Theory and Operation e-learning courses available from Agilent education

 

 