HI Lars,
Yes, you can run the test on a VI. Specific to your system, a failing result could indicate a leak within the inlet or at any of the in-line connections of your GSV, so the results may not be clear without a point leak detector. I would also…
I have a 6490 QQQ. Each time I run either autotune or check tune, I get masses that are out of tolerance especially in MS2. The system seems to pass if the Dynamic Ramp Table enable button is checked. This is associated with the collision cell and detector…
Hello, H2 is the fuel for the flame in the FID, so if you are out of H2 then the FID will not be able to light. That is very likely the reason why you are not seeing a signal from the FID.
I’m sorry for the delayed reply to your message. Do you still need assistance with this? If you do, please take into consideration the notes below. However, if the issue was resolved, could you please share what the resolution was?
From what I understand…
Based on your latest results, I would say that nothing is fundamentally wrong with the optical unit in terms of noise, so I believe this is more an applicational problem - your passed quick noise test points to that direction. The error 1050 with the…
No, the flows on the two columns downstream of the switch have to be equal. They are feed by the same pressure source, so since MSD is under vacuum and the FID is ambient, the column connected to the MSD has to have more restriction, to keep the flows…
I have decades of experience with consumer products, all on Agilent GC, GCMS, and HPLC.
If I was needing to quantitate 1,2-octanediol in cosmetics samples, I'd extract the sample with dimethylformamide and add BSTFA to make trimethylsilyl derivative…
Disconnect the sample probe from the valve and start the leak and restriction test. If it passes this step, it means that there is some restriction in the sample probe, enough to cause that part of the test to fail. Try a brand new probe. If it fails…
Seems then to indicate something in the flow path when including the mixer and damper. First, make sure that you do an elasticity calibration because you have changed important parts in the flow path. Then re-run the pump leak rate test with the mixer…
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power on the Agilent ICP-MS Foreline Pump when fully powering on the Agilent ICP-MS System, or after performing maintenance.
Background
The ICP...
This Information Applies To: Agilent ICP-MS MassHunter Software version 5.1 and above
Issue
The user manuals for Agilent ICP-MS Systems for MassHunter 5.1 and above are included in the ICP-MS MassHunter Help and Learning Center. There are manuals...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Elevated oxide ratio values are defined as greater than 1.5% for x-lens configuration or greater than 3% for s-lens configuration. Possible causes are contamination...
This Information Applies To: Agilent GC/MS Systems
Issue
The software version is a valuable information usually required for troubleshooting, software updates preparation or to check installed options.
Table of contents
MassHunter GC/MS Acquisition...
This Information Applies To: Agilent LC modules with RFID tag readers
Issue
Agilent uses RFID (radio frequency identification tag) tag readers in some HPLC components such as detector lamps and flow cells, pump heads, purge valve and mixers. However...
This Information Applies To: Agilent LC systems with Lab Advisor software.
Issue
In Lab Advisor, the Logs and Results register is available for gathering history and additional information on the system. It presents data collected from the configured...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
What information does the ICP-MS instrument status indicator provide?
Background
The ICP-MS Instrument Status Indicator on the instrument, and in the...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power off Agilent ICP-MS systems before maintenance, for long-term shutdown, or for power cycling the instrument. Power cycling may be required when...
This Information Applies To: Agilent OpenLab CDS
Issue
Steps to close an instrument connection to a primary instrument controller, to start a successful connection to a secondary instrument controller, or to configure the instrument.
Background...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or replacement of consumable...