So ive had a very high sample with Methanol (around 8000 mg/l) and now i still have contamination around 500 mg/l in my blanks. Ive tried injecting IPA and replaced the syringe. Any other tips? Ive also pumped up the oven/column tempature to a…
5 psi inlet pressure running hydrogen with your configuration leads to 2.3 ml/min to the MS and 1.02 ml/min to the FID.
You are welcome. Are you using a two-hole ferrule in the injection port or a splitter of some type? The two-holed ferrules are..…
Canadian_Chemist ,
There is an example of printing text instead of the final concentration based on the LOQ outlier in this discussion.
(+) Report builder - Final Concentration Expression Change - Forum - Mass Spectrometry Software - Agilent Community…
Yes, it is off on all gasses, always by roughly 20% (occasionally nearly 30% or as low as 15%, depending on the instrument). Air, Hydrogen, Makeup (N2), carrier (H2 or He) on every instrument (12 GC and 3 GC/MS, which haven't been evaluated). It's also…
Hello,
3-way valve is a good idea, but unfortunately all I had was several 5180-4160 Tee, 1/8 in, brass, and several 0100-2144 Ball valve, 1/8 inch.
So I decided to make following setup.
In the setup below I am using He as carrier gas and N2 as…
When trying to close a connection on an instrument with an active single run or sequence the OpenLAB CDS software will usually give a message that it will close connection when the runs are finished. Today when trying to close 2 instruments on the same…
Hello Agilent Users,
Im missing a feature in Openlab since day 1. How do I override and start a run if the GC is not ready ?
For example: the FID temperature is set at 250°C in the method and the actual temperature is at 220°C (it is still heating…
Hi,
My baseline is around 8pA in standby state. H2 is coming from a tank and I can rule out contaminations in the gas source because the gas line is splitted with T piece: one is going to the FID EPC and one is going to the inlet PCM. H2 is the carrier…
Hi all
Apparently the engineer who did the mainboard replacement made a mistake by plugging the MCGV cable to the SSV cable, this made the valve failed to switch (EE 2040.2) and electronic fuse of SSV (EE2049.0) error always appear. It was out of my…
This Information Applies To: HES ion source in 5977B/C GC/MSD, 7010 triple quadrupole GC/MS systems
Issue
The filaments are replaced as part of routine maintenance, or when they break (zero output current).
Items Required
Gloves, nylon,...
This Information Applies To: Agilent 5977A/B/C GC/MSD
Issue
The Electron Multiplier Horn must be replaced when it is contaminated or the EMV value reaches 3000 V with a clean ion source.
Items Required
Electron multiplier horn (5977B/C:...
This Information Applies To: Agilent 5973/ 5975/5977 GC/MSD with rotary vane foreline pump
Issue:
The oil mist filter is recommended to be replaced annually. If the color of the filter has changed or there is smell from the filter, the oil mist...
This Information Applies To: Agilent EI Source
Issue:
During normal operation of the GC/MS, the EI SST/Inert ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filaments...
This Information Applies To: Agilent G7129A/B/C Vialsampler
Issue:
Replacing the needle and needle seat of the Vialsampler is required when:
The needle is visibly damaged.
Leaks or blockages in the vialsampler are observed, especially between...
This Information Applies To: Agilent G1311A/B/C, G7111A/B, Quaternary Pumps, and G1310A/B, G7110B Isocratic Pumps, G1312A/B/C, G7112B Binary Pumps
Issue:
How to diagnose no flow or no pressure issues on binary, quaternary, or isocratic pumps.
...
This Information Applies To: Agilent UV and RID detectors
Issue
A clogged flow cell causes high backpressure. The clog can be removed by flushing the system with solvents at high pressure. If you suspect a clog or experience high backpressure due...
This Information Applies To: Agilent G1312X and G7112X binary pumps.
Issue:
A blockage before the pressure sensor, e.g. in a pump head, will not give a high pressure or overpressure system error. Instead, the pump may show one of various error messages...
This Information Applies To: Agilent 1100, 1200, 1260 LC Pumps.
Issue:
The following procedure allows the cleaning of residue from the pump's check valves, often caused by polymerization of aged acetonitrile or salt-based solutions that may prevent...
This Information Applies To: Agilent LC Systems
Issue
After analyzing dirty or concentrated samples, some sample compounds may retain in the column and other parts of the HPLC system. If the column re-equilibration after each run is not enough to...