Hello,
So, you should be able to add another column to your report table and change the value to = Compound_Concentration * Sample_Amount / 100 to remove the sample amount and percentage. If you need to remove a multiplier or dilution you will need…
Hello,
LIMS ID is not available as a token for sample name, datafile name, or in CDS DA for report output. You can see the options in the screenshot below. It can be a value in the CSV file but not used in the name. The only other option would be…
Our G4220A (1290 Bin Pump) is setup to work with the seal wash currently. It operates with our instrument with no issue. If I go into the instrument configuration and check the box to enable the pump seal wash (The pump is installed), it will lose connection…
Hi RDP0314,
One thing to check is the injection settings. This behavior is common when the GC is expecting a series of manual injections. To check, click on the Instrument menu shown on your screen capture above and 'Select Injection Source' or similar…
The GPC/SEC Columns Selector Tools help you navigate to the best column choice for your application. Guidance can be by mobile phase, application, manufacturer’s part number cross-reference, USP designation, or EP monograph.
Hi, I always use P/A factor adjustment under Acq/Setup. That works very well. In all your calibration samples a p/a factor adjustment is done. Takes only a few seconds extra.
We do not maintain listings of spectra in the ST Japan libraries. Please contact ST Japan to obtain a listing of all the spectra in the library. They can reached at Contact - S.T.Japan Europe GmbH (stjapan.de)
When using the Microlab PC SW in conjunction with the 21CFR11 software, the PUBLIC database must be part of the data PATH. It needs to be added to the GROUP (easiest to do it here), the PROJECT or the USER. Otherwise the error described will always be…
This Information Applies To: HES ion source in 5977B/C GC/MSD, 7010 triple quadrupole GC/MS systems
Issue
The filaments are replaced as part of routine maintenance, or when they break (zero output current).
Items Required
Gloves, nylon,...
This Information Applies To: Agilent 5977A/B/C GC/MSD
Issue
The Electron Multiplier Horn must be replaced when it is contaminated or the EMV value reaches 3000 V with a clean ion source.
Items Required
Electron multiplier horn (5977B/C:...
This Information Applies To: Agilent 5973/ 5975/5977 GC/MSD with rotary vane foreline pump
Issue:
The oil mist filter is recommended to be replaced annually. If the color of the filter has changed or there is smell from the filter, the oil mist...
This Information Applies To: Agilent EI Source
Issue:
During normal operation of the GC/MS, the EI SST/Inert ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filaments...
This Information Applies To: Agilent G7129A/B/C Vialsampler
Issue:
Replacing the needle and needle seat of the Vialsampler is required when:
The needle is visibly damaged.
Leaks or blockages in the vialsampler are observed, especially between...
This Information Applies To: Agilent G1311A/B/C, G7111A/B, Quaternary Pumps, and G1310A/B, G7110B Isocratic Pumps, G1312A/B/C, G7112B Binary Pumps
Issue:
How to diagnose no flow or no pressure issues on binary, quaternary, or isocratic pumps.
...
This Information Applies To: Agilent UV and RID detectors
Issue
A clogged flow cell causes high backpressure. The clog can be removed by flushing the system with solvents at high pressure. If you suspect a clog or experience high backpressure due...
This Information Applies To: Agilent G1312X and G7112X binary pumps.
Issue:
A blockage before the pressure sensor, e.g. in a pump head, will not give a high pressure or overpressure system error. Instead, the pump may show one of various error messages...
This Information Applies To: Agilent 1100, 1200, 1260 LC Pumps.
Issue:
The following procedure allows the cleaning of residue from the pump's check valves, often caused by polymerization of aged acetonitrile or salt-based solutions that may prevent...
This Information Applies To: Agilent LC Systems
Issue
After analyzing dirty or concentrated samples, some sample compounds may retain in the column and other parts of the HPLC system. If the column re-equilibration after each run is not enough to...