The GC inlet must have 20 ml/min or more total flow for good inlet pressure/flow control. If the total flow is too low, air can actually diffuse into the system and the oxygen will damage the column.
Using an Agilent Headspace Sampler the samples are…
Hello ljones031 ,
I would suggest copying the installation files to a folder on the local hard drive and then disconnecting any external drives before attempting installation. Most MassHunter programs will try to install to a D: drive it if is found…
Hello,
If you need the formatting of a table, it can be done that way. I do not like to use multiple hidden tables for that kind of report as they can be hard for later customization. The example below uses a single hidden table and custom code to…
Hello Agilent Users,
Im missing a feature in Openlab since day 1. How do I override and start a run if the GC is not ready ?
For example: the FID temperature is set at 250°C in the method and the actual temperature is at 220°C (it is still heating…
Hello Marty,
thanks for your reply. I tested your hint. It worked and I'm confident to code the formula in the CC.
Thank you very much and have a nice weekend,
Sarah
Hello well ,
I know this is an old forum post, but just wanted to add to what ISH was saying.
The answer still stands, 6890 is not compatible with the 7650 ALS. The GC models it is compatible with are (at the time of this comment at least):
- 8890…
Hi,
I passed this query on to a colleague who was involved with developing these columns, this was their response:
"It might be worth checking if they are filtering their mobile-phase and samples as it's possible the guard may be fouled. They could…
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue
The Agilent 2100 Bioanalyzer system run ends prematurely and the following error appears in the logbook: Run aborted on port x. Instrument error occurred on port x, Unusual...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue
The Agilent 2100 Bioanalyzer system electropherograms in the Agilent 2100 Expert software have a wavy, uneven baseline ( Figure 1 ).
Figure 1. Wavy, uneven baseline on Bioanalyzer...
This Information Applies To: Agilent AriaMx systems.
Issue
When IP addresses of an AriaMx instrument are reassigned, it will appear under the new and the previous IP address in the AriaMx Instrument Explorer. This situation can be prevented using...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Sometimes, for support purposes, it is necessary to send the diagnostic test result file to Agilent support representatives.
This file contains detailed information about the...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx Real-Time PCR software
Issue
How to change the discovery protocol port in the AriaMx software.
Steps to Follow
The AriaMx software uses the UDP Broadcast protocol for...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
The instrument diagnostic tests should be run whenever it is suspected that there could be a fault with the instrument hardware.
Background
The instrument diagnostics consist...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Error 45 is displayed on the AriaMx instrument touch screen.
Background
Error 45 is shown when the temperature recorded by one or more thermistors is outside the expected...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx showing a series of errors after delivery, guide for troubleshooting and resolving temperature-related error.
Background
When an AriaMx is delivered, it could take...
How to Access Manual Tune in GC/MS MassHunter and MSD Productivity ChemStation
Multiple Results
This Information Applies To: MassHunter acquisition software for GC/MS systems and MSD Productivity ChemStation for all Agilent GC/MS systems
Issue
Manual tuning lets you interactively set the MSD parameters, such as lens voltages and tuning masses...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
How to find the AriaMx software and firmware version that you have installed.
Steps to Follow
There are two different ways to find out AriaMx software version:
In the...