Hi, the peak area has no influence on the calculation of Mw when measuring with concentration detectors only (RI, UV, ELSD e.g.). When performing Viscometry & Lightscattering measurements, the peak area is needed for the calculation. As an additional…
Occasionally the tune files get corrupted and you need to go back to an old one or start again from the default atune.u file. Glad the problem is resolved!
Part 1 included many questions and some answers about the maximum abundance that should be detected at the electron multiplier. Another important consideration is how much sample should go into the column.
See Which Electron Ionization Ion Source? Part…
Thank you for taking a moment to review the Community Guidelines. You agree to all of the guidelines posted and linked here when using the Agilent Community. These guidelines are here to help you be a valuable contributor to the community and make sure…
Hi, I have an old 6890 with a jet direct card. We lost power to the instrument and since then it has not been able to connect to the PC. I tried updating the bootp software with the correct mac address and power cycled instrument with the same issue.…
Support and educational resources for the Microplate Centrifuge and Automated Centrifuge Loader .
Automation Solutions Knowledge Base
Microplate-Handling Device Guides
PDF user guides
Support Documents
G5582A Centrifuge Installation Checklist…
-Have you tried a diferent column ? An old column could have retention time issues and peak distortion.. -Have you had any pressure fluctuations during your run ? Try running a Pump leak rate test, to see if the problem is not in the pump-head. -Do a…
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...