A tiny bit of fluff on the HED or the hard wire that goes from the HED feedthrough to the HED can dance with -10,000 V on it. The first easy thing to try is to vent, turn off, remove the two front cables that go to the analyzer, open the analyzer door…
I asked this same question to the Agilent tech who was here for our quarterly maintenance. You won't find any documentation and they won't give you any. Apparently, you can brick your instrument if you don't know what you're doing. The scripts are more…
Hi,
you should set your instrument to "user tune" in startup. But plasma correction is intended for a regualr setup. you can try to run it with your "special configuration".
Under "Startup Configuration" change the Ignition Sequence to "user tune…
This is the correct Tune Solution. It should be fine if you dilute it 1:10.
From the information i have, i cannot say why the res/axis is failing.
Best Regards Chris
That's great. From the Home screen do the following:
Click the <Advanced Features> button
Click the <Library Management> button
At the top of this dialog, click on the library in question and see if the spectra are displayed on the bottom of the…
KbNYSP1701 Is the system functioning properly now?
How many runs on this system?
8890 and Intuvo: Service Mode, Instrument, Usage
7890: Service Mode , Diagnostics, Instrument Status
6890: Options, Diagnostics, Instrument Status
For reference, see:
(+) GCMS Ion Source Cleaning - Videos - GC/MS Portal - Agilent Community
(+) How to clean an Agilent GCMS Ion Source - Files - GC/MS - Agilent Community
(+) Apart. Together. Two Minutes! GCMS Extractor Ion Source - Videos - GC…
This Information Applies To: 4200, 4150 and 2200 TapeStation, TapeStation analysis software A02.02 or higher.
Issue:
When using the Agilent 4200, 4150 and 2200 TapeStation systems, the DV 200 of FFPE RNA samples can be calculated within the TapeStation...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, ProSize Data Analysis Software
Issue
How to export a ladder as a size calibration file and import it into a different run in ProSize.
Background
...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
It is possible to set up the 2100 Expert software so that the files are automatically exported each time a chip run is complete.
Steps to Follow
To turn...
This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software
Issue:
How to identify and troubleshoot late migration of the ladder, marker, and sample peaks in data generated on the Agilent 2100 Bioanalyzer system.
Background
...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
If communication between the 2100 Bioanalyzer and the PC cannot be established, or if an intermittent communication loss occurs, follow the instructions in...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
Instructions on how to import from or export data into LIMS software.
Steps to Follow
Data import
When importing sample names and sample comments from...
This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software version B.02.08 or later
Issue
When printing a report to PDF or HTML with the 2100 Expert software, the error "Failed to Generate Report" appears.
Background
This error...
This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software
Issue
Resetting upper or lower markers from their automatically assigned settings in the Agilent 2100 Expert software.
Background
The Agilent 2100 Expert software...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert Software
Issue
Samples with a broad signal such as NGS smears should be analyzed by setting up a region. Analyzing a region is also termed smear analysis.
Steps to Follow...