This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software
Issue
Resetting upper or lower markers from their automatically assigned settings in the Agilent 2100 Expert software.
Background
The Agilent 2100 Expert software will assign the markers automatically. The software is able to assign the markers as they each have a known size (bp/nt/kDa) and therefore should migrate within a predetermined time frame. If a marker has not migrated within the expected time frame, or if there are other peaks close to the marker, a different peak may be incorrectly assigned as the marker. Incorrect marker peak assignment may result in inaccurate sizing and/or quantitation.
Steps to Follow
To check which peaks should be assigned as a marker:
If you find that a marker peak has been set incorrectly:
Figure 1. Assigning a peak as a marker in the 2100 Expert software
Learn how to effectively operate, maintain, and troubleshoot your Agilent 2100 Bioanalyzer system:
Agilent 2100 Bioanalyzer System Expert Software User's Guide on Agilent.com Agilent 2100 Bioanalyzer System Maintenance and Troubleshooting Guide on Agilent.com |
For Research Use Only. Not for use in diagnostic procedures.