This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
If communication between the 2100 Bioanalyzer and the PC cannot be established, or if an intermittent communication loss occurs, follow the instructions in this article to re-establish a stable instrument connection.
Steps to Follow
Instrument cannot connect or error: Instrument connection timeout
Figure 1. COM Port drop-down menu within the 2100 Expert software
Intermittent loss of communication or error
Intermittent loss of communication may occur with one of the following errors:
|
Caution: Changing Windows Settings on the PC may affect the functionality of other software installed on the PC.
|
Intermittent loss of communication and error Unexpected firmware termination
This error is a hint of a potential hardware defect. Create a support package and contact Agilent support.
Learn how to effectively maintain and troubleshoot your Bioanalyzer: Agilent 2100 Bioanalyzer System Maintenance and Troubleshooting Guide on Agilent.com Agilent 2100 Bioanalyzer System 2100 Expert Software User's Guide on Agilent.com |
For Research Use Only. Not for use in diagnostic procedures.