Hello, I am having troubles with the MS40+ pump I use for my ICP-MS machine. I did an oil change yesterday and cleaned out the inside as well as the filters (the oil was pretty low since I had forgotten to check it for a while) and now when I go to turn…
Hello, I am having troubles with the MS40+ pump I use for my ICP-MS machine. I did an oil change yesterday and cleaned out the inside as well as the filters (the oil was pretty low since I had forgotten to check it for a while) and now when I go to turn…
Hi mubashirraza, PCBs are not commonly sampled via headspace. SPE is more commonly used as a cleanup step followed by standard auto-sampler injection. PCBs are more often detected using MS or uECD due to their more selective operation. I've attached some…
Hello
I recommend to replace X-belt.
P/N 5410047500 Belt timing – 880 mm
When initializing if the belt tension is not enough, It could be failed.
Please check it.
Thanks! To add, I just ran the software verification tool and the MassHunter GCMSAcq_base.xml report failed. File Report Summary
Missing files or invalid files found:
Invalid Files :1
No system file difference found
Invalid Files …
Hi there! We have a Cary 60. We are doing preventive maintenance, but the noise test fails and it took up to three or four times to get a successful result. What can we do to improve this and get a passed result the first time?
We have several 1220 Infinity II system (G4290B) on our location. They are connected same type of computers with same windows. Only one of them has connection issue as below. Since there is connection to detector (I can control detector fully), I am not…
As per my knowledge after changing any component in GC (FID assembly,Injector liner parts,Column) the instrument need to be kept for conditioning about minimum of one hour other wise as you did three to four blank injections has to run. Due to the temperature…
This Information Applies To: Agilent 4200 TapeStation 4150 TapeStation, TapeStation controller software
Issue
The system verification can be run to check the TapeStation performance, for example, when you suspect an instrument hardware fault.
...
This Information Applies To: Agilent 4200 and 4150 TapeStation software.
Issue
How to uninstall and reinstall the 4200 and 4150 TapeStation software.
Background
An uninstallation and reinstallation of the TapeStation software should be performed...
This Information Applies To: 2200 TapeStation instruments, 2200 TapeStation controller software
Issue:
The Agilent 2200 TapeStation controller software won't launch but displays the correct serial number in the top-right corner of the loading screen...
This Information Applies To: 4200, 4150 and 2200 TapeStation, TapeStation analysis software A02.02 or higher.
Issue:
When using the Agilent 4200, 4150 and 2200 TapeStation systems, the DV 200 of FFPE RNA samples can be calculated within the TapeStation...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, ProSize Data Analysis Software
Issue
How to export a ladder as a size calibration file and import it into a different run in ProSize.
Background
...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
It is possible to set up the 2100 Expert software so that the files are automatically exported each time a chip run is complete.
Steps to Follow
To turn...
This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software
Issue:
How to identify and troubleshoot late migration of the ladder, marker, and sample peaks in data generated on the Agilent 2100 Bioanalyzer system.
Background
...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
If communication between the 2100 Bioanalyzer and the PC cannot be established, or if an intermittent communication loss occurs, follow the instructions in...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
Instructions on how to import from or export data into LIMS software.
Steps to Follow
Data import
When importing sample names and sample comments from...