Hello
I have a problem with a SPS3 during initializing, It appears LEDs 2 and 4 ON, when it had to move along X axis. I have change all boards, cables and encoders wheels. Do you know what it means LED 2 and 4 ON????
Hello
I have a problem with a SPS3 during initializing, It appears LEDs 2 and 4 ON, when it had to move along X axis. I have change all boards, cables and encoders wheels. Do you know what it means LED 2 and 4 ON????
Hello
I recommend to replace X-belt.
P/N 5410047500 Belt timing – 880 mm
When initializing if the belt tension is not enough, It could be failed.
Please check it.
Hello
I recommend to replace X-belt.
P/N 5410047500 Belt timing – 880 mm
When initializing if the belt tension is not enough, It could be failed.
Please check it.