Hello, H2 is the fuel for the flame in the FID, so if you are out of H2 then the FID will not be able to light. That is very likely the reason why you are not seeing a signal from the FID.
Seems then to indicate something in the flow path when including the mixer and damper. First, make sure that you do an elasticity calibration because you have changed important parts in the flow path. Then re-run the pump leak rate test with the mixer…
Disconnect the sample probe from the valve and start the leak and restriction test. If it passes this step, it means that there is some restriction in the sample probe, enough to cause that part of the test to fail. Try a brand new probe. If it fails…
Hello experts!! I have 6460 LCMSMS system. It unfortunately failed in tuning even in brand new tuning solution. Auto tune repeatedly failed in peak width adjustment in MS1. What could be the problems and possible answers? Cleaning is done up to ion transfer…
No, the flows on the two columns downstream of the switch have to be equal. They are feed by the same pressure source, so since MSD is under vacuum and the FID is ambient, the column connected to the MSD has to have more restriction, to keep the flows…
I have decades of experience with consumer products, all on Agilent GC, GCMS, and HPLC.
If I was needing to quantitate 1,2-octanediol in cosmetics samples, I'd extract the sample with dimethylformamide and add BSTFA to make trimethylsilyl derivative…
I'm trying to create a report with a column which calculates the percentage of all integrated peaks above a specific threshold. I have created a report parameter to format the number of significant digits, this is a request of some of our clients. At…
I found the issue. I have "Hide duplicates" selected which prevents any additional Fails from being displayed. However, if I uncheck that box, there is a Fail in every row. This is a summary result, is there a way to make it display only once per group…
Hello everyone,
We have a problem with the acquisition methods on GC, but not on HPLC.
When we want to change an acquisition method that is in "approved" status to "generic" status in order to modify it, it is impossible to do so because the box is…
This Information Applies To: Agilent OpenLab ChemStation Edition Software (C.01.xx)
Issue
When your Agilent OpenLab ChemStation Edition software fails to launch with errors " Login failed with the following message: Configuration system failed to...
This Information Applies To: Agilent OpenLab ChemStation Edition Software (C.01.xx)
Issue
Agilent OpenLab ChemStation Edition software fails to launch with error containing the words " Autostart Macro failed ”. This article will provide you some...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer and Oligo Pro II instruments, and their respective controller software.
Issue Failed to build pressure error comes up shortly after starting a separation...
This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems equipped with a gate valve.
Issue
Removing the Ion Transfer Capillary for maintenance on instruments with gate valve. The gate valve allows ion capillary removal and...
This Information Applies To: Agilent OpenLab ChemStation, Agilent OpenLab CDS, Agilent OpenLab EZChrom
Issue
Due to the license error “ An error occurred connecting to the license server: The desired vendor daemon is down. Check the Imgrd log file...
This Information Applies To: HPLC Advisor app
Issue
Legacy HPLC method needs to be translated to a new column or LC system (HPLC to UHPLC or Analytical to Preparative LC).
Requirements
Install the most recent version of the HPLC Advisor...
This Information Applies To: Agilent 7890 A/B, 8860 and 8890 GC Systems with Split/Splitless (SSL), Multimode (MMI) or Programmable Temperature Vaporization (PTV) Inlets
Issue:
The Inlet Leak & Restriction Test is executed after every inlet maintenance...
This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7, ICP Expert version 7.X
Issue
The AVS can contribute to instrument performance issues such as sensitivity and precision. During troubleshooting, bypassing (or skipping...
This Information Applies To: Agilent GC Systems
Issue
Loss of peak resolution can be a combination of decrease in separation and increased peak width. Loss of resolution can impact on analytical reproducibility and accuracy.
Other chromatographic...
This Information Applies To: Agilent GC Systems
Issue
Contamination that creates ghost and carry over peaks can be introduced at different points including sample preparation, introduction, and separation. The additional peaks can cause false positive...