Hello blairchall ,
I’m sorry for the delayed reply to your message. Do you still need assistance with this? If you do, please take into consideration the notes below. However, if the issue was resolved, could you please share what the resolution was…
We use Agilent NGA Analyzer GC based on 4 column configuration. Hayesep Q pre column, Hayesep Analytical column, molsieve column for TCD and one capillary column for FID. The H2S elution takes place in the Hayesep analytical column. But now a days our…
H ztarbs,
The air would not be for an inlet supply. This is most likely a detector support gas for FID. Check your method to make sure that the FID is not on in a standby method. FID air consumption is 400 to 450ml/min when on.
Regards
James
Hello, all you need to do in the method is to uncheck each of the FID flows, and uncheck the electrometer and uncheck the Flame on. With everything unchecked for the FID settings the software will ignore your FID. Under the Signals section you will also…
Thank you for your response Today when you open the GC we noted the increase returned valid but not connected and the Remote LED off we tried to configuration instrument but a message appears(failed to connect to 6890 at 100.100.100.101)but the GC IP…
Hi again
try the last hope ,
first try to select your reporting method in ( DA report template ) , the sequence accept it but i don't know if it will success to print ,
you may try it
the second thing if the first failed ,then try to edit your…
I want to check Total hydrocarbon by FID without identifying the hydrocarbon species . Agilent suggest to give me DB-1 15m, 0.32mm, 0.10u as a non polar column to elute all HC without separation as one peak . Does this column will work for Total hydrocarbon…
Hi silencer2008,
the 1 µL/s rule means the injection speed used for splitless injections, which is very slow compared to split injection where you want to inject as fast as possible.
Two interesting articles would be:
(+) Pre-Columns - the forgotten…
We've been having issues with our 7890B gas chromatograph. We have it connected to a workstation running Windows 7 SP1 and OpenLab/Chemstation version C.01.07 [27]. We have a Gerstel MPS attached as a headspace sampler.
We have been trying to run…
How to Create a New Database in Spectroscopy Database Administrator (SDA)
Multiple Results
This Information Applies To: Spectroscopy Database Administrator (SDA)
Issue
If the old database is full, or after a specific time interval, it is necessary to create a database in SDA. If you receive a message telling you that the database is full...
Update Database Group Path in Spectroscopy Configuration Manager (SCM)
Multiple Results
This Information Applies To: Agilent Spectroscopy Configuration Manager (SCM)
Issue
It is necessary to add a database to a group path in SCM after creating a new database in Spectroscopy Database Administrator (SDA) . If access to an older database...
How to Access Spectroscopy Database Administrator Software (SDA)
Multiple Results
This Information Applies To: Agilent Spectroscopy Database Administrator Software
Issue
The SDA application can only be accessed by a Windows user who is a member of the Windows Administrators group and Windows SDAAdministrators group.
Steps...
This Information Applies To: AriaMx real-time qPCR system
Issue
The instrument fails to reach the required temperature during a run.
Resolution
Check if the temperature of the environment around the instrument is too low resulting in the instrument...
This Information Applies To : Agilent ELSDs G4260A, G4260B, G4261A, G4621B, G7102A, G7102B
Content :
If the evaporator tube of ELSD becomes contaminated with nonvolatile material resulting in poor chromatography, please use the procedure in this...
This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900)
Issue
The oil should be replaced at least once every six months, or when the oil color is indicates it needs to be changed . The foreline pump should be checked monthly. This article...
This Information Applies To: Agilent ICP-MS MassHunter versions 4.4 - 4.6 and 5.x.
Issue
After performing an autotune, the batch tune report will not generate or print.
Background
ICP-MS MassHunter uses an application called ICP-MS Data Analysis...
This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems equipped with a gate valve
Issue
Installing the Ion Transfer Capillary after cleaning or replacement in Agilent LC/MS instruments with a gate valve.
Required Items...
This Information Applies To: Agilent OpenLab CDS 2.x
Issue
This article outlines the procedure on how to quantify uncalibrated compounds in OpenLab CDS 2.x. using reference compounds in the processing method compound table.
Background
Uncalibrated...
This Information Applies To: Agilent OpenLab CDS 2.x
Issue
This article outlines the procedure on how to quantify the uncalibrated compounds in OpenLab CDS 2.x when the analytical method requires a fixed response factor.
Background
Uncalibrated...