This Information Applies To: Agilent Spectroscopy Configuration Manager (SCM)
Issue
It is necessary to add a database to a group path in SCM after creating a new database in Spectroscopy Database Administrator (SDA). If access to an older database is no longer needed by the group, remove the database on the Server and Database Selection page. This will keep users from saving data to the incorrect database location. If archiving the database, remove the database from the group path in SCM before detaching it in SDA.
Requirements
Any of the following software products must be installed: SpectrAA CFR, ICP Expert CFR, WinUV CFR, and MicroLab PC CFR.
Steps to Follow
Figure 3. Spectroscopy Configuration Manager
1. Groups icon, 2. Group to modify, 3. Edit...
Figure 5. Server and Database Selection
1. Server, 2. New database, 3. >> button
Figure 6. Server and Database Selection
1. New database added, 2. OK button
Click Yes in the Create Directory dialog box (Figure 7).
Figure 7. Create Directory dialog box
Tips: It is possible for a Group to have access to multiple attached databases at a time. If no other changes to the group path are needed, skip to step 4.
If access to an older database is no longer needed by the group, remove the database on the Server and Database Selection page. This will keep users from saving data to the incorrect database location. If archiving the database, remove the database from the group path in SCM before detaching it in SDA. See step 3 for instructions to remove a database from the group path. |
Figure 8. Server and Database Selection
1. Database to remove, 2. << button, 3. OK button
Caution: Do not move, rename, or delete a database without first detaching the database in SDA. Failure to detach first will result in software errors, making the software unusable, and will need to be resolved by an Agilent Service Representative.
|