Hello isoterma, Maybe try a new jet ? Cleaning doesn't always get rid of restriction in the jet. I see your supply is supplying both detectors but if the pressure is on the lower side of acceptable raising the pressure may cure the issue. For the 6890…
Hello,
I will assume you mean you are using the backup restore tool for CDS 2.4 not the ECM XT import scheduler. If you have setup an import scheduler task to back the ECM-XT system up that would be self-referential and would not actually back anything…
Hello, I will take a picture when I go back to the lab tomorrow. I must also let you know that we were able to extract an old method from the GC. This method had been downloaded into the GC from before everything crashed. When using this method, the GC…
After servicing pump, test still fails. We are using n-heptane as solvent/carrier, is correct to use n-heptane for the pressure/leak test? Or, do we need to do it with isopropanol?
I have seen cases where the rough pump shuts down because it's working too hard, which will cause the MSD to shut down also. Typically the MSD will have an error message of "Vac Sys Fail" or something similar.
Sounds good. Just wanted to double check why it was not working. I had a noise and drift test fail before on LabAdvisor and it was due to a test not functioning properly. The Leak test and Pressure test both passed and system has been working fine. Thanks…
Hello mr.white, You can do this with a sequence summary report template: Use 'Multi Signal Plot Overlaid' as you've done already; in the properties, the signals need to be overlaid. You may group by signal e.g. FID Front signal/ FID Back signal or, if…
Do you have the Polyboost on? If not, the wavelengths below 190nm will fail. It would be good to know which wavelengths are failing. On the wavecal tab, click "Store dark current" and see where the wavelength calibration file is being saved. Then using…
Csiquet_flex2000 my guess for your 4 signal sources the Save check box is unchecked, so you don't see them as an online signal during acquisition nor as a saved raw data file. The Signal settings under the GC Parameters is only for GC based detectors…
This Information Applies To: Agilent 8890/8860 GC Systems with SSL, MMI, PTV, or VI inlets.
Issue
Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure...
This Information Applies To: Agilent 6000 Series LC/MS systems equipped with MS40+ rotary vane pumps
Issue
Instructions on how to replace the filter in an Agilent MS 40+ pump as part of annual pump maintenance
Background
The filter in the...
This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900) with ICP Expert 7.5.x Software
Issue
The Recommended Routine Maintenance Schedule does not adjust for the number of samples analyses or the sample matrix, both of which...
This Information Applies To: Agilent 7890/7820 GC Systems with SSL, MMI, PTV, or VI inlets.
Issue
Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure...
This Information Applies To: Agilent LC pumps
Issue:
Stuck or obstructed check valves (Active Inlet Valve (AIV), Passive Inlet Valve (PIV), Outlet Ball Valve (OBV)) can cause issues such as no-flow or no-pressure, moderate to strong drops in pressure...
This Information Applies To: Agilent 1260 Infinity II multisampler, 1290 Infinity II multisampler, 1260 Infinity II SFC multisampler, and 1260 Infinity II bio-Inert multisampler
Issue
The needle bends while moving to the wash port or showing one...
This Information Applies To: Agilent OpenLab CDS 2.x
Issue
This article outlines the step-by-step instruction on how to overlay chromatograms for OpenLab CDS 2.x
Steps to Follow
Load data from result sets or single samples. See Load or add...
This Information Applies To: Agilent Lab Advisor Software, Agilent Multicolumn Thermostat
Issue:
Agilent Lab Advisor is a great tool to diagnose the issue of your HPLC. Running the tests that are relevant to your specific issue in the Service &...
This Information Applies To: Agilent OpenLab CDS (ver 2.x)
Content:
The following procedure shows how to review completed injections in a running sequence in OpenLab CDS (ver 2.x).
Steps to Follow:
Select the sequence from the Run Queue...