This Information Applies To: Agilent 7890/7820 GC Systems with SSL, MMI, PTV, or VI inlets.
Issue
Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure is too high, or the split vent flow cannot be reached.
The replacement frequency will vary with the GC application.
For other GC models, see: Replacing the Split Vent Trap Cartridge Filter on a 8890/8860 GC System
Items Required
Steps to follow
Warning: The oven or inlet may be hot enough to cause burns. If either is hot, wear heat-resistant gloves to protect your hands.
|
Warning: The split vent trap may contain residual amounts of any samples or other chemicals you have injected into the GC. Follow your company’s safety procedures for handling these types of substances while replacing the trap filter cartridge.
|
Figure 1 Trap assembly.
1. Split vent valve, 2. Knurled nut on weldment assembly, 3. Retaining clip, 4. Split vent trap screw
Caution: Be careful not to stress (twist or bend) the tubing between the split vent valve and the trap as it make cause small fractures and leaks.
|
Tips: This procedure is also available in the video: Change the Split Vent Trap - GC Troubleshooting Series
|
Learn how to effectively maintain your 7890 GC:
GC-7890-2232z - How to Replace the 7890 Split Vent Trap e-learning course available from Agilent education.
|