Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
Greetings Agilent Community,
As with many of the errors I encounter with my system, these two happened after I foolishly left the lab to pursue a life outside the world of LCMS.
(1) From the instrument log, the first error read "Failed to run acquisition…
When attempting to submit a sequence in EZChrom, it fails to submit and returns this error message:
This happens each time a sequence is submitted. Any help would be appreciated!
Hi Agilent community.
I'm a LC-MS equipment user and I had a problem with the MassHunter Acquisition B 08.01 software.
When I try to open the Acquisition software, appears the software logo but immediately appears an error dialog box. This error dialog…
Hi everyone,
We have a 6546 QTOF, the following error keeps popping up during acquisition; Failed to start logbook server. A shared violation has occurred. How can I resolve this?. I'll appreciate your help. Thanks
This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900) with ICP Expert 7.5.x Software
Issue
The Recommended Routine Maintenance Schedule does not adjust for the number of samples analyses or the sample matrix, both of which...
This Information Applies To: Agilent 7890/7820 GC Systems with SSL, MMI, PTV, or VI inlets.
Issue
Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure...
This Information Applies To: Agilent LC pumps
Issue:
Stuck or obstructed check valves (Active Inlet Valve (AIV), Passive Inlet Valve (PIV), Outlet Ball Valve (OBV)) can cause issues such as no-flow or no-pressure, moderate to strong drops in pressure...
This Information Applies To: Agilent 1260 Infinity II multisampler, 1290 Infinity II multisampler, 1260 Infinity II SFC multisampler, and 1260 Infinity II bio-Inert multisampler
Issue
The needle bends while moving to the wash port or showing one...
This Information Applies To: Agilent OpenLab CDS 2.x
Issue
This article outlines the step-by-step instruction on how to overlay chromatograms for OpenLab CDS 2.x
Steps to Follow
Load data from result sets or single samples. See Load or add...
This Information Applies To: Agilent Lab Advisor Software, Agilent Multicolumn Thermostat
Issue:
Agilent Lab Advisor is a great tool to diagnose the issue of your HPLC. Running the tests that are relevant to your specific issue in the Service &...
This Information Applies To: Agilent OpenLab CDS (ver 2.x)
Content:
The following procedure shows how to review completed injections in a running sequence in OpenLab CDS (ver 2.x).
Steps to Follow:
Select the sequence from the Run Queue...
This Information Applies To: Agilent Liquid Chromatography (LC) systems
Issue:
Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column.
Steps to follow:
Store the...
This Information Applies To: Agilent G7129A/B Autosampler with Sample Thermostat/Cooler G7167-60101 or G7167-60005.
Issue:
The autosampler may stop working from time to time, potentially giving off loud noises and switching off the light. The only...