A tiny bit of fluff on the HED or the hard wire that goes from the HED feedthrough to the HED can dance with -10,000 V on it. The first easy thing to try is to vent, turn off, remove the two front cables that go to the analyzer, open the analyzer door…
I asked this same question to the Agilent tech who was here for our quarterly maintenance. You won't find any documentation and they won't give you any. Apparently, you can brick your instrument if you don't know what you're doing. The scripts are more…
Hi,
you should set your instrument to "user tune" in startup. But plasma correction is intended for a regualr setup. you can try to run it with your "special configuration".
Under "Startup Configuration" change the Ignition Sequence to "user tune…
This is the correct Tune Solution. It should be fine if you dilute it 1:10.
From the information i have, i cannot say why the res/axis is failing.
Best Regards Chris
That's great. From the Home screen do the following:
Click the <Advanced Features> button
Click the <Library Management> button
At the top of this dialog, click on the library in question and see if the spectra are displayed on the bottom of the…
KbNYSP1701 Is the system functioning properly now?
How many runs on this system?
8890 and Intuvo: Service Mode, Instrument, Usage
7890: Service Mode , Diagnostics, Instrument Status
6890: Options, Diagnostics, Instrument Status
For reference, see:
(+) GCMS Ion Source Cleaning - Videos - GC/MS Portal - Agilent Community
(+) How to clean an Agilent GCMS Ion Source - Files - GC/MS - Agilent Community
(+) Apart. Together. Two Minutes! GCMS Extractor Ion Source - Videos - GC…
This Information Applies To: Agilent ICP Expert software.
Issue
The software version is often asked and is a valuable information: to see if any updates are needed, if certain options are available etc.
Steps to Follow
Launch ICP Expert...
This Information Applies To: Agilent OpenLab CDS
Issue
This article provides step-by-step instructions to calculate peak-to-valley ratio in OpenLab CDS.
Background
The peak-to-valley ratio is calculated to indicate the quality of peak separation...
This Information Applies To: Agilent 6500 Series Q-TOF LC/MS systems, MassHunter Workstation 10.1 or below and MassHunter Workstation 11.0 .
Issue
Fully reset the Q-TOF LC/MS system via a power cycle without venting. For users who experience slow...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, and Oligo Pro II systems
Issue
The volume of sample used by Agilent parallel capillary electrophoresis systems during a run.
Resolution
The...
This Information Applies To: Agilent 2100 Expert software, 2100 Expert Security Pack, 2100 Bioanalyzer
Issue
When opening the 2100 Expert software and using the Security Pack, the error message " Start of launcher application failed - The requested...
This Information Applies To: Agilent 6890, 6850, 7890, 8890, 8860, Intuvo 9000 GC systems.
Issue
The exchange of the ignitor in a Flame Photometric Detector (FPD) may be required in the following scenarios:
Preventive or periodic maintenance...
This Information Applies To: Agilent 1290 Infinity Quaternary Pump G4204A, 1290 Infinity II Flexible Pumps G7104A/C, and G7131A/C.
Issue
If the flexible/quaternary pump does not generate flow or build up pressure, it could be due to a stuck inlet...
This Information Applies To: Agilent 5200, 5300 and 5400 Fragment Analyzer systems
Issue
Separation times for the Fragment Analyzer for each array format and kit.
Resolution
Caution : The times in this article are the length of time...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, Oligo Pro II, ProteoAnalyzer systems
Issue
Guidance for installing the Agilent parallel capillary electrophoresis instrument controller and/or...
This Information Applies To: OpenLab CDS (version 2.x) system
Issue:
This article explains how to export OpenLab CDS 2.x data as Archived file.
For copying OpenLab CDS 2.x data to another Projects or another OpenLab CDS systems, you can use data...